Impact of External Factors on the Accuracy of Infrared Thermographs for Measuring Elevated Body Temperature

Authors:
Poster Author(s)
Mazdeyasna, Siavash, FDA/CDRH; Wang, Quanzeng, FDA/CDRH
Center:
Contributing Office
Center for Devices and Radiological Health

Abstract

Poster Abstract

Background:

Infrared thermographs (IRTs) have often been used for identifying febrile individuals in public during pandemics. Standardized approaches for performing IRT-based fever screening have been described in a recent document (ISO/TR13154). However, the scientific underpinnings of some aspects of this document remain unclear, as do the role of several potential confounding factors.

Objective:

The purpose of this study is to investigate the effects of external factors – including environmental conditions (e.g., ambient temperature, ambient relative humidity (RH)), and deployment parameters (e.g., working distance, viewing angle, setting temperature of external temperature reference source (ETRS)) – on the accuracy of IRTs for elevated body temperature (EBT) detection.

Method:

The influence of the ETRS setting temperature on IRT accuracy was investigated in the 30-40˚C range with a calibration source (CS). The effects of ambient temperature, RH, working distance, and viewing angle were theoretically evaluated through bench tests by setting the ETRS and CS temperature at 37 ˚C. Computer simulations were conducted to demonstrate the influence of the environmental factors on the total atmospheric transmission, total energy/radiosity received by the IRT, and the temperature readout.

Results:

Results shows that ETRS setting temperature of 35˚C to 37˚C is the optimal for EBT screening. The accuracy of the temperature measurement decreased with increasing viewing angle. The error was less than 0.05˚C for viewing angle lower than 30-dgree for both IRTs. The influence of the ambient temperature, RH, and working distance within the ranges of 18-32˚C, 15-80%, and 0.4-2.8 m respectively were investigated. The bench tests show differences up to 0.97˚C and 0.11˚C without and with ETRS respectively. Considering working distance, ambient temperature, and RH in the ranges of 0.2-3 m, 15-35˚C, 5-95% respectively, the computer simulations show miscalculation less than 0.30 ˚C and 0.04 ˚C without and with ETRS respectively.

Conclusion:

Our research into the performance evaluation of the IRT systems has provided significant insights toward the design of the least burdensome standardized test methods. We conducted a series of benchtop measurements and computer simulations to address external factors relevant to the effectiveness and practical implementation of IRTs for EBT screening.


Poster Image
Impact of External Factors on the Accuracy of Infrared Thermographs for Measuring Elevated Body Temperature

Download the Poster (PDF; 1.59 MB)