FDA Questions for the Circulatory
System Devices Panel
January 13, 2005
P040043
W.L. Gore & Associates, Inc.
GORE TAG Thoracic Endoprosthesis
Safety
Pivotal Study (TAG 99-01): The primary
safety endpoint for the pivotal study was the proportion of subjects who
experienced ≥1 major adverse event (MAE) through 1 year
post-treatment. Comparisons were made
between subjects treated with the original design of the GORE TAG Thoracic
Endoprosthesis and open surgical repair.
The safety null hypothesis
was that the proportion of subjects who experienced ≥1 major adverse
event (MAE) through 1 year post-treatment was equal in the Control subjects and
the TAG subjects. The alternate hypothesis
was that the proportion of subjects who experienced ≥1 major adverse
event (MAE) through 1 year post-treatment was less in the TAG subjects than in
the Control subjects. The primary
safety endpoint is a composite outcome consisting of the occurrence within 1
year post-procedure of any of the following MAEs:
|
|
|
Results: The
proportion of subjects who experienced ≥1 MAE through 1 year post-treatment
was significantly lower (p < 0.001) in the TAG (42%) vs. Control (77%)
group. Ten (7.1%) TAG subjects had no
12-month follow-up visit. Assuming that
all 10 TAG subjects experienced a MAE through 1 year post-treatment, the
estimated 1-year MAE incidence in the TAG group increased from the 42% to 49%. However, the significance level for the comparison with the
Control group remained < 0.001.
Confirmatory Study (TAG
03-03): The primary safety endpoint for the confirmatory study
was the proportion of subjects who experienced ≥1 major adverse event
(MAE) through 30 days post-treatment.
Comparisons were made between subjects treated with the GORE TAG
Thoracic Endoprosthesis (TAG) and open surgical repair (Control).
The safety null hypothesis
was that the proportion of subjects who experienced ≥1 major adverse
event (MAE) through 30 days post-treatment was equal in the Control subjects
and the TAG subjects. The alternate hypothesis
was that the proportion of subjects who experienced ≥1 major adverse
event (MAE) through 30 days post-treatment was less in the TAG subjects than in
the Control subjects. The primary
safety endpoint is a composite outcome consisting of the occurrence within 30
days post-procedure of any of the MAEs as defined in TAG 99-01 and listed
above.
Results: The proportion of
subjects that experienced ≥1 MAE through 30 days post-treatment was
significantly less (p < 0.001) in 03-03 TAG subjects (12%) compared to 99-01
Control subjects (70%).
Panel Question 1:
Please comment on whether the results
of the clinical studies with the above mentioned safety endpoints provide
reasonable assurance of safety for the current device design in the intended
population.
Effectiveness
Pivotal Study (TAG 99-01):
The efficacy null hypothesis was that the proportion of
subjects treated with the TAG device free from a major device-related event
through the 12-month follow-up visit would be <0.8. The alternate hypothesis was that the
proportion of subjects treated with the TAG device free from a major
device-related event through the 12-month follow-up visit would be >0.8. This endpoint was a composite outcome
consisting of subjects who were free from the following major device-related
events: aneurysm enlargement, endoleaks, aneurysm rupture, branch vessel
occlusion, deployment failure, extrusion/erosion, lumen obstruction, prothesis
material failure, prosthesis migration, and prosthesis realignment.
When formalizing the
efficacy hypotheses, the efficacy of open surgical repair was assumed to be
100%. A point estimate of 80% was
judged to be a reasonable efficacy outcome for the endovascular treatment. The Agency and the sponsor agreed to an
analysis plan where the device would need to show superior safety as the
efficacy was expected to be less than that for surgical repair.
Results: The freedom
from a major device-related event for the TAG device was 87 to 94 percent. Eight (6%) subjects experienced ≥1
major device-related event through the 12-month follow-up visit. The efficacy null hypothesis was that the
proportion of subjects free from a major device-related event through the
12-month follow-up visit would be < 0.8.
The resulting efficacy estimate was 0.94 with a lower bound 95%
confidence interval of 0.90, and the null hypothesis was rejected (p <
0.0001) in favor of the alternate hypothesis that the probability was
≥0.8. These calculations assume
that the 10 TAG subjects without a 12-month follow-up visit had no major
device-related events. If one assumes
that all 10 subjects without a 12-month follow-up visit had a major
device-related event in this period, the estimate of the probability of not
having a major device-related event was 0.87, and the null hypothesis was still
rejected (p = 0.02, 1-sided).
Confirmatory
Study (TAG 03-03): The primary efficacy endpoint was the proportion of
subjects treated with the GORE TAG Thoracic Endoprosthesis who were free from a
major device-related event through the 30 day follow-up visit. Device-related events in the TAG group were
presented descriptively.
Results: There were no major device-related events (no endoleaks, access
failures, occlusion of branch vessel, prothesis migration, or aneurysm
enlargement) through the 30-day follow-up visit in 03-03 TAG subjects compared
to 6 (4%) subjects in the 99-01 TAG cohort.
There were 6 (12%) subjects with minor endoleaks in the 03-03 TAG
subjects, compared to 19 (14%) reported for the 99-01 TAG subjects. One subject with a minor proximal endoleak
(noted 1-month procedure) and a distal endoleak (noted 92 days-post-procedure)
underwent placement of 2 additional Gore TAG Endoprostheses 141 days post-procedure.
Panel Question 2: Please comment on whether the results of the
clinical studies, with the above mentioned endpoints provides reasonable
assurance of effectiveness for the current device design in the intended
population.
Patient
Comparability
In the pivotal study there was a
statistically higher prevalence of symptomatic aneurysms in the open surgical repair control
group as compared to the GORE TAG
Thoracic Endoprosthesis (38% control, 21% TAG).
Panel Question 3: Please
comment on whether the difference in the prevalence
of symptomatic aneurysms is clinically
significant and whether this affects your comments from questions 1 and 2.
Labeling
One
aspect of the pre-market evaluation of a new product is the review of its
labeling. The labeling must indicate which patients are appropriate for
treatment, identify potential adverse events with the use of the device, and
explain how the product should be used to maximize clinical benefit and
minimize adverse events. If you recommend approval of the device, please
address the following questions regarding product labeling.
Panel Question 4:
The proposed INDICATION FOR USE
for this device is as follows: Endovascular repair of aneurysm of the
descending thoracic aorta. Please
comment on whether the indication for use adequately defines the patient
population studied and for which the device will be marketed. Please address the need to include the
required anatomical parameters for this device in the indications for use
statement. (Note: As a point of reference, the indications for
use of AAA approved endovascular grafts are attached as Appendix 1 to this
document.)
Panel Question 5:
Based on the clinical
investigation experience, please comment on whether there are any additional warnings, precautions, or
contraindications that you think should be included, either specific to this
device or from a generic standpoint for endovascular grafts intended to treat
thoracic aneurysms.
Panel Question 6:
Please
provide any additional comments you have on the labeling.
Panel
Question 7: Please comment on the adequacy of the proposed
physician training plan, as described in the panel package (Appendix E).
Post-Market
Study
Panel
Question 8: Please comment on the adequacy of the proposed
post-approval study plan, as described in the panel package (Appendix F).
Appendix 1
Indications for Use for Currently Approved
Endovascular Grafts for Treatment of Abdominal Aortic Aneurysms
Cook, Inc.
The Zenith AAA Endovascular Graft with the H&L-B One-Shotä Introduction System and ancillary components is indicated for the endovascular treatment of patients with abdominal aortic, aortoiliac or iliac aneurysms having morphology suitable for endovascular repair, including:
· An abdominal aortic aneurysm with a diameter ³4 cm
· Aortic or aorto-iliac aneurysm with a history of growth ³0.5 cm per year
W.L. Gore & Associates, Inc.
The EXCLUDER Bifurcated Endoprosthesis is intended to exclude the aneurysm from the blood circulation in patients diagnosed with infrarenal abdominal aortic aneurysms (AAA) disease and who have appropriate anatomy as described below:
• adequate iliac/femoral access;
• infrarenal aortic neck treatment diameter range of 19-26 mm and a minimum aortic neck length of 15 mm;
• proximal aortic neck angulation <60°; and
• iliac artery treatment diameter range of 8-13.5 mm and iliac distal vessel seal zone length of at least 10 mm.
Guidant
The
ANCURE Tube System is indicated for the endovascular treatment of infrarenal
abdominal aortic aneurysms (AAA) in patients having:
• adequate iliac/femoral access;
• infrarenal non-aneurysmal neck length of at
least 15 mm and a diameter of no greater than 26 mm;
• distal segment neck length of 12 mm and
diameter of no greater than 26 mm; and
• morphology suitable for endovascular repair.
The ANCURE Bifurcated System is indicated for the endovascular treatment of infrarenal abdominal aortic or aorto-iliac aneurysms in patients having:
• adequate iliac/femoral access;
• infrarenal non-aneurysmal neck length of at
least 15 mm and a diameter of no greater than 26 mm;
• distal segment lengths of at least 20 mm and
diameters no greater than 13.4 mm; and
• morphology suitable for endovascular repair.
abdominal aortic or aorto-iliac aneurysms in patients whose anatomy does not allow the use of a tube or bifurcated device and having:
• adequate iliac/femoral access;
• infrarenal non-aneurysmal neck length of at
least 15 mm and a diameter of no greater than 26 mm;
• one distal segment length of at least 20 mm and
diameters no greater than 13.4 mm; and
• morphology suitable for endovascular repair.
Medtronic
The AneuRx Stent Graft System is indicated for the endovascular treatment of infrarenal abdominal aortic or aorto-iliac aneurysms having:
• adequate iliac/femoral access;
• infrarenal, non-aneurysmal,
neck length of greater than 1 cm at the proximal and distal ends of the
aneurysm and an inner vessel diameter 10-20% smaller than the labeled device
diameter;
• morphology suitable for endovascular repair; and
• one of the following:
-
aneurysm diameter >5
cm;
-
aneurysm diameter of 4-5
cm and which has also increased in size by 0.5 cm in the last 6 months; or
- aneurysm which is twice the diameter of the normal infrarenal aorta.