Indication Adjunctive therapy to whole brain radiation for the treatment of brain metastasis originating from breast cancer
Study RT-009 is a randomized, multi-center, open-label study designed to demonstrate the added benefit of RSR13 to whole brain radiation therapy (WBRT) in the treatment of brain metastasis. Patients on Treatment Arm A received an initial 100 or 75 mg/kg dose of RSR13 over 30 minutes via a central venous access device. Whole brain radiation therapy was given within 30 minutes after the completion of the RSR13 infusion. Patients on Treatment Arm B received WBRT without a placebo. WBRT was given as 3cGy per day for 10 days in both treatment arms. Supplemental oxygen was given to both treatment arms. The subjects were stratified prior to randomization into one of the four strata: RPA Class I, RPA Class II and NSCLC primary, RPA Class II and breast primary, and RPA Class II and ‘Other’ primary tumors. The primary efficacy endpoint was overall survival. The applicant is seeking the following indication based on efficacy results in a non-prespecified subgroup of 115 patients with primary breast cancer in study RT-009:
The following table summarizes the FDA’s analysis of efficacy based on the updated survival data (January 2004 cut-off) from study RT-009:
|
|
Overall WBRT (N=267) |
Overall RSR13 + WBRT (N=271) |
NSCLC/Breast 10 Subgroup WBRT (N=206) |
NSCLC/Breast 10 Subgroup RSR13 + WBRT (N=208) |
Breast 10 Subgroup WBRT (N=55) |
Breast 10 Subgroup RSR13 + WBRT (N=60) |
|
Median Survival in
months (95% C.I.) |
4.5 (3.7, 5.4) |
5.3 (4.5, 6.2) |
4.5 (3.8, 5.4) |
5.9 (4.7, 7.0) |
4.6 (3.8, 6.2) |
8.8 (6.0, 11.3) |
|
Hazard Ratio (95%
C.I.) |
0.87 (0.73, 1.04) |
0.84 (0.68, 1.03) |
0.63 (0.422, 0.946) |
|||
|
P-value* (log-rank test) |
0.13 |
0.084 |
0.02 |
|||
* Not adjusted for multiple analyses
The following table summarizes the FDA’s analysis of the most common grade 3 and 4 adverse events from study RT-009:
RT-009: Grade 3 and 4 Adverse Events
|
|
Control Total patients =264 N (%) |
RSR13 Total patients=268 N (%) |
|
Hypoxia |
6 (2) |
30 (11) |
|
Dyspnea |
19 (7) |
8 (3) |
|
Headache |
11 (4) |
18 (7) |
|
Pneumonia |
7 (3) |
14 (5) |
|
Hypotension |
1 (<1) |
5 (2) |
|
Acute renal failure |
1 (<1) |
6 (2) |
|
Cerebral edema |
1 (<1) |
4 (1) |
There were some imbalances between the two treatment arms in the subgroup of patients with primary breast cancer. Most notably, these included baseline characteristics such as number of brain lesions. In addition, imbalances related to therapy during or subsequent to radiation included duration of oxygen administration and subsequent therapy. These imbalances favored the RSR13 + WBRT treatment arm. It is possible that these imbalances, in particular imbalance in tumor load, influenced the results in the breast subgroup.
Questions to the Committee