Questions for Consideration
P040001
- Patients
who had the X STOP implanted had a higher incidence of musculoskeletal
events, including lower back disorders, lower extremity disorders, hip
disorders, upper back disorders, and neurological and neuropathological
disorders, compared to the control group.
Although these adverse events were considered by the sponsor to be
not device related, changes in spinal biomechanical function that occur
with the limits to extension could also be a potential source of pain. The sponsor provides a report of a pre-clinical
study which characterizes the effects of the device in cadaveric specimens,
showing an increase in canal and foraminal dimensions at the implanted
level in the extended position, with no change in the dimensions at the adjacent levels. Please discuss the clinical significance
of the musculoskeletal and other adverse events seen in the trial, and discuss
whether the effects of the device on surrounding segments or on spinal
biomechanics have been adequately addressed.
- Based
on your knowledge of the biomechanics of the spine and the nature of
spinal stenosis, please discuss whether there is a clinical basis for
pooling the outcomes of the one- and two-level patients.
- The
device labeling states that this device limits extension. In the pre-clinical cadaveric studies,
ranges of flexion-extension were recorded under measured applied
loads. The clinical radiographic
measurements, however, were performed on static plain radiographs. Please discuss the interpretation of
the measurements made on the clinical patients’ radiographs, as it relates
to device effectiveness.
- Fewer
than 50% in the X STOP treated group and fewer than 5% in the control group
achieved overall successful outcome.
These results are considerably lower than what had been predicted
at the outset of the study. In
this study, an operative treatment was compared to a nonoperative
treatment in patients who had already failed conservative treatment,
including epidural injections. A
majority of patients had had symptom duration for more than 2 years prior
to entering the study. Patients in
both groups went on to have more than one epidural injection and/or laminectomies. In 10-15% of the X STOP treated
patients who improved, symptoms returned during the course of the
study. Moreover, there was a trend
toward different results for use of this device at one vs two levels.
(a) Based on the data from this study, please
discuss the appropriate population who might benefit from this device.
(b) Given the
historical success rates for laminectomy, please discuss what impact the
effectiveness results of this study have in relation to our interpretation of
the risks and benefits of treatment with the X STOP device.
- In
this study, the protocol did not define what criteria were to be used in
either group to determine when or whether patients proceeded to
laminectomy; it also did not define whether to administer additional
epidural injections to patients in the control group. Some patients in the investigational (X
STOP) group received the control treatment (epidural injection for pain)
rather than proceeding to laminectomy, and it is not clear whether success
in those patients was due to temporary relief from the injection or to the
X STOP. Please describe the potential
impact on the interpretation of the study result of these confounding
factors.
- Under
CFR 860.7(d)(1) , safety is defined as reasonable assurance, based on
valid scientific evidence, that the probable benefits to health under
conditions of the intended use, when accompanied by adequate directions
for use and warnings against unsafe use, outweigh any probable risks. Do the clinical data in the PMA provide
reasonable assurance that the device is safe?
- Under
CFR 860.7(e)(1) effectiveness is defined as reasonable assurance that , in a significant portion of the
population, the use of the device for its intended uses and conditions of
use, when accompanied by adequate directions for use and warnings against
unsafe use, will provide clinically significant results. Do the clinical data in the PMA provide
reasonable assurance that the device is effective?