Draft Panel Questions
1.
Preclinical
animal evaluations (included
in this PMA or in the panel
package? Otherwise we should not refer
to data that they do not have) have shown that
the rate and amount of DMSO can cause vasospasm and vascular wall damage. Additional preclinical animal evaluations included in the PMA package demonstrated
that long term chronic inflammatory responses will occur in response to the
presence of the embolization agent. Patients
undergoing staged embolization procedures for Cerebral Arteriovenous
Malformations will be exposed repeatedly to the potential for
to
DMSO -mediated
vessel damagesince they will undergo staged
procedures over a period of time prior to resection. If the patient’s AVM
is not resected, vessel wall damage due either to repeated-direct DMSO
cytotoxicity or to an inflammatory response to the embolic agent could result
in angionecrosis with resulting patient morbidity. Repeat instillation of the product, as would occur in staged
embolization procedures, has not been evaluated in animal preclinical
models. Do you believe that
the preclinical data in the PMA adequately support the safety of repeated exposure to DMSO as required in the proposed use of the product? If not, please provide suggestions on the
additional preclinical studies that you believe are needed to demonstrate the safety of
the repeated exposure to DMSO.-Do you believe
additional animal evaluation should be conducted to more completely assess for
repeat-DMSO vessel wall exposure and potential adverse effects? Do you have
any recommendations regarding the amount of DMSO a patient should be exposed to
over a 24 hour period or the length of time between embolization procedures?
1.
1.
1.
21 CFR
860.7(d)(1) states that there is a reasonable assurance that a device is safe
when it can be determined that the probable
benefits to health from use of the device for its intended uses, when
accompanied by adequate instructions for use and warnings against unsafe use,
outweigh any probable risks. The clinical study determined that there were
slightly more total complications with the subject device than with the control
device. Of particular concern were the device-related complications of
difficulty in removing the catheter (10 – Onyx, 0 – nBCA), catheter shaft
rupture (2 – Onyx, 0 – nBCA) and poor penetration/visualization (5 – Onyx, 0 –
nBCA). In addition, there were more
deaths and strokes (both 2 – Onyx, 0 – nBCA) in the treatment group than in the
control group. Please discuss whether
the safety data for Onyx Liquid Embolic System provides a reasonable assurance
that it is safe for this use with respect to the approved control device.?
Effectiveness Question
1.
The device is
intended for presurgical embolization and therefore, the material is meant to
be removed during surgical resection of the AVM. Although patients were enrolled in this study based upon the
criterion that they were surgical candidates, in some cases, the clinical
course of treatment changed so that it was not in few patient’s best interest
to such that some subjects did not undergo
surgical excision, post-embolization.
With regard to the potential for this product to be used in pediatric
patients, and in recognition that there is very limited long-term implantation
safety information for the product, Considering that it is probable that this scenario will also arise under clinical use, if you
recommend approval, do you believe a long-term
follow-up study for patients not undergoing surgical resection of their AVM
should be conducted as a post-approval commitment?
2. 21
CFR 860.7(d)(1) states that there is a reasonable assurance that a device is
safe when it can be determined that the probable benefits to health from use of
the device for its intended uses, when accompanied by adequate instructions for
use and warnings against unsafe use, outweigh any probable risks. Please
discuss whether the data in the PMA for Onyx Liquid Embolic System provide a
reasonable assurance of
safety.
3. 21
CFR 860.7(e)(1) states that there is a reasonable assurance that a device is
effective when it can be determined, based upon valid scientific evidence, that
in a significant portion of the target population, the use of the device for
its intended uses and conditions of use, when accompanied by adequate
directions for use and warning against unsafe use, will provide clinically
significant results. Please
discuss whether the data in the PMA for the Onyx Liquid Embolic System provide a reasonable assurance of
effectiveness.
1.
4. A number of complications were observed related to the placement of the devicein the study that appear to relate to user training. Of particular concern were the device-related
complications of difficulty in removing the catheter (10 – Onyx, 0 – nBCA),
catheter shaft rupture (2 – Onyx, 0 – nBCA) and poor penetration/visualization
(5 – Onyx, 0 – nBCA). Please comment on the sponsor’s proposed training
plan and whether you believe it is adequate to help ensure proper device placement. (Note: we
should send the training plan in the second mailing, if we have not yet sent
it)use.
5. The
device is intended for presurgical embolization and therefore, the material is
meant to be removed during surgical resection of the AVM. Although patients were enrolled in this
study based upon the criterion that they were surgical candidates, in some
cases, the clinical course of treatment changed such that some subjects did not
undergo surgical excision, post-embolization.
Considering that it is probable that this scenario will also arise under
clinical use, if you recommend approval, do you believe a long-term follow-up
study for patients not undergoing surgical resection of their AVM should be
conducted as a post-approval commitment?