Microsulis
Microwave Endometrial Ablation (MEA™) System
(P020031)
DRAFT Discussion Questions
Safety and Effectiveness
1.
Using
an intent-to-treat analysis, the following represent the pivotal study results:
|
|
MEA™ |
REA |
|
Success Rate |
87% (187/215) |
83.2% (89/107) |
|
Amenorrhea Rate |
55.3% (119/215) |
45.8% (49/107) |
Does the panel agree that these results demonstrate the
clinical effectiveness of the MEA™ System?
2.
Does
the Panel agree with the sponsor's analysis of the 27 cases of serious adverse
events that occurred with the MEA system in commercial use outside of the
US? In particular, regarding the
serious injuries that occurred with uterine perforation, does the Panel agree
that these cases were the result of relative thinning of the uterine wall (due
to trauma or surgical history), inappropriate pre-treatment, and failure to
follow the Instructions for Use?
3.
In commercial
use (outside the U.S.), this device has been associated with transmural full
thickness burns. Microsulis has
provided their analysis of the injuries and the contributing factors. The company has also provided an analysis of
the changes made to the commercial labeling and training to minimize this risk,
for example:
a.
requirement
for proctoring for new users;
b.
no
mechanical D&C;
c.
use of
ultrasound on all patients to determine uterine wall thickness; and
d.
perform
hysteroscopy on all patients after cervical dilation and prior to insertion of
the MEA™ applicator to confirm that the uterine cavity is intact.
The training and labeling items identified in items a-d were
included in the pivotal trial (n=215) without report of serious adverse
event. (Many of these items were also
practiced at three commercial centers (n=1400) without report of serious
adverse event and over 2,400 consecutive patients have been treated worldwide
without report of incident since November 2002.)
Given the detailed information on the serious adverse events
observed in past commercial use and the sponsor’s analysis of the contributing
factors, does the panel believe that the measures taken by the sponsor to
improve the training and labeling will
sufficiently reduce or eliminate the risk of full thickness burns?
4.
The sponsor
is proposing a minimum uterine thickness of 10 mm as measured by
ultrasound. Microsulis believes that
the maximum depth of tissue destruction with the MEA System is 7 mm. What does the panel consider to be a
reasonable minimal uterine wall thickness to prevent full thickness burns
considering:
a.
the
variability in uterine perfusion among the treatment population and the
associated impact on depth of penetration
b.
the
uncertainty of the temperature at which thermal damage occurs; and
c.
the
imprecision in ultrasound measurements.
5.
Does
the panel agree with the instructions provided in the labeling for an
ultrasound evaluation in 3-views?
a.
Are
the instructions adequate for performance of 3-view pre-operative ultrasound?
b.
What
is the appropriate level of training and/or experience necessary for the person
performing the ultrasound evaluation?
c.
What
is the appropriate timing for the ultrasound evaluation?
6.
As
with any endometrial ablation system, operation in the presence of a uterine
perforation is associated with significant morbidity. The intra-operative
safety features included as part of the MEA procedure for detecting a uterine
perforation include:
a.
hysteroscopy
after cervical dilation but prior to insertion of the MEA™ applicator;
b.
comparison
of length of applicator to uterine sound measurement; and
c.
software
feature (temperature rise gate (TRG)) that detects placement of the applicator
outside of the uterine cavity at the initiation of treatment.
Are these methods sufficient for identifying a uterine
perforation prior to treatment?
Labeling
& Training
7.
Does
the panel have any comments on the labeling and training provided by the
sponsor? Are there specific
recommendations related to the proposed:
a.
Indications;
b.
Contraindications;
c.
Warnings;
and
d.
Precautions?
Post-market
Study
8.
Under
current FDA guidance, patients from the pivotal study are scheduled to be
followed for a total of 3 years after the procedure (1 year pre-market and
2-years post-market). If the panel
votes to recommend approval of the MEA™ System, is there a need for additional
post-approval studies? If so, what is
the purpose of such studies and what are the key elements of the study
design?
Note: Post-approval
studies may provide additional information about an approved device; however,
the safety and effectiveness must be demonstrated before approval. The results of a post-approval study should
not be expected to change the “approval” status of the device.