Circulatory
System Devices Panel Questions
May
29, 2003
P020039
Cardima,
Inc.
Revelation™
Tx and NavAblator™ Ablation Catheter System
Study Design and Execution
The
sponsor conducted a single-arm, multi-center study. The study protocol identified three linear ablation lesions to be
performed with an optional fourth lesion.
The Indications for Use statement is based on the procedure dictated in
the study protocol.
1. The sponsor presented
acute procedural data on 116 patients, all of whom had the posterolateral and
posteroseptal linear lesions performed with the Revelation Tx catheter. The tricuspid isthmus lesion was made with a
variety of catheters throughout the two phases of the study (refer to the table
below, retrieved from Table A-19 of submission).
Devices used to create
tricuspid isthmus lesion
|
Catheters |
Number of patients (%) |
||
|
Phase IIb n=33 |
Phase III n=75 |
Total N=108 |
|
|
Revelation Tx only |
14 (42.4) |
7 (9.3) |
21 (19.4) |
|
Revelation Tx,
other |
7 (21.2) |
1 (1.3) |
8 (7.4) |
|
Revelation Tx,
NavAblator |
n/a |
2 (2.7) |
2 (1.9) |
|
NavAblator only |
n/a |
47 (62.7) |
47 (43.5) |
|
NavAblator, other |
n/a |
10 (13.3) |
10 (9.3) |
|
Other only |
12 (36.4) |
8 (10.7) |
20 (18.5) |
a. Please discuss how the multiple catheter combinations
affect the conclusions that may be drawn from this study.
b. Please discuss the ability to analyze the device
outcomes versus treatment outcomes in this study. In particular, can you comment on whether
the safety and effectiveness results for this study may be attributable to one
specific catheter? Do the treatment strategies
employed in the study support the proposed indications for use statement?
2. The study identified the
acute procedural effectiveness endpoints as demonstration of either reduction
in the amplitude, fragmentation or widening of local electrograms or appearance
of split potentials. These endpoints
were not consistently measured and/or recorded in the data collection forms. The sponsor indicated in the PMA submission
that no conclusion can be made about the acute procedural success endpoint. It is unknown what measurement or
electrogram characteristic was used by the individual investigator to determine
when the linear lesion was completed.
Therefore, each investigator may have performed a slightly different
ablation procedure.
a. Please discuss how the
lack of a measurable procedural endpoint affects data analysis for this
clinical trial.
b. Please discuss whether the
study provides sufficient information to instruct the user of the catheter
system as to procedural goals or endpoints when treating an individual patient.
a.
It
is unknown if each transmission represented a discrete atrial fibrillation
episode. Given that the patients knew
that a certain amount of episodes were required to be admitted into the study,
please discuss the potential problems with accuracy in the counting of episodes
at baseline and at follow up.
b.
During
the 6th month post procedure, the protocol required patients to
transmit weekly recordings and, in addition, whenever they had symptoms of
atrial fibrillation. Fifty three out of
83 patients (63.9%) transmitted fewer than four times in the 6th
month; thus, 36.1% of patients completely complied with the protocol
requirement. Please discuss how incomplete
compliance with transmissions of rhythm strips impacts measurement of the
primary effectiveness endpoint.
Safety
4. The safety endpoint was
listed as incidence of adverse events over 24 months and major complications
that occurred in the first 7 days post procedure. A total of 5/116 (4.3%, with the upper limit of a 95%
confidence interval being 9.4%) patients had a major complication
within a week of the right atrial ablation procedure. If all patients who required pacing within two weeks are
included, the major complication rate would be 6.9% (8/116), with the upper limit
of a 95% confidence interval being 13.0%. Please comment on
whether the results of the clinical study provide reasonable assurance of
safety for the intended use.
5. A total of 20/88 (22.7%)
patients had a permanent pacemaker implanted during the long-term follow-up
period. Nine of the 20 also had an AV
node ablation procedure. In this
patient population, please comment on whether this rate of permanent pacemaker
implantation represents a significant safety concern for the device. Given the lack of a control group, please
comment on how one would determine an acceptable rate of permanent pacemaker
implantation.
Effectiveness
6. The primary
effectiveness hypothesis was “subjects with ≥5 episodes during the 30 day
baseline period, a reduction of 50% or more is required to constitute a
clinical success. For those subjects
with 3-4 episodes during the baseline period, a reduction of 75% is required to
constitute a clinical success.” These reductions were to occur in patients
either on the same medication regimen or decreased dosage. The results of the clinical study have been
described in the presentation.
a.
Do
the clinical data provide a reasonable assurance of effectiveness of the system?
b.
Please
discuss the clinical utility of the primary endpoint of a percent decrease in
atrial fibrillation episodes, as opposed to a cure for paroxysmal atrial
fibrillation. Does a significant
decrease in PAF episodes constitute adequate evidence for effectiveness?
7.
A
secondary effectiveness endpoint of the study was improvement in quality of
life. Given the potential bias
introduced with a non-randomized unblinded study, please comment on the device
system's demonstration of improvement in quality of life.
8.
If
you believe that additional data are necessary to demonstrate reasonable
assurance of safety and effectiveness of the Cardima ablation system, please
address the following questions:
a. Please clarify if additional analyses on the
current data set may be performed to provide adequate information to support
safety and effectiveness.
b. Please comment if the collection of
additional data using the current patient selection criteria and outcome
measures would be adequate to support safety and effectiveness.
c. Please comment if a new prospective trial is
needed to provide adequate information to support safety and
effectiveness.
Labeling
9.
Labeling
for a new device should indicate which patients are appropriate for treatment,
should identify potential device-related adverse events, and should explain how
the device should be used to optimize its risk/benefit profile. If you recommend device approval, please
address the following:
a. Does the Indications for Use, as stated
below, adequately define the patient population and procedural use for which
the device will be marketed?
"The
Cardimaâ Inc., REVELATION®
Tx Microcatheter Ablation System is indicated for treatment of Atrial
Fibrillation in patients with drug refractory paroxysmal atrial fibrillation by
mapping, pacing, and ablating with a set of continuous linear lesions in the
right atrium.
b. Based on the study results, please discuss
whether the proposed warnings, precautions, and contraindications are
acceptable.
c. Please discuss whether the instructions for
use adequately describe how the device should be used.