P010018
QUESTIONS FOR PANEL DISCUSSION
1.
What are the concerns regarding the incidence of induced
cylinder with significant axis shift, and its consequent effect on efficacy?
2. Is 12-month follow-up sufficient to provide
reasonable assurance of safety and efficacy?
There are 21 eyes available at 24 months. Should data for these eyes also be required in the labeling?
2.
Does the refractive correction obtained with this
device, in light of the rate of change of mean MRSE over time and the incidence
of over- and undercorrection, justify the potential risks?
4. Are there concerns regarding the increased
incidence of visual symptoms from pre-op levels?
5.
Do the safety and efficacy data presented in this PMA
support approval of this device for the requested indication? Is the requested indication appropriate as
worded, based on the study outcome?
6.
What are your recommendations for labeling regarding regression of effect,
induction of cylinder, and incidence of visual symptoms? Are there additional labeling
recommendations?