News 03/12/1992
P92-7 Food and Drug Administration
FOR IMMEDIATE RELEASE Sharon Snider - (301) 443-3285
The Food and Drug Administration has asked Shiley Inc. of Irvine,
Calif., the maker of Bjork-Shiley heart valves, to notify patients and
physicians that risk of fracture for some sizes of these valves may be
higher than previously thought.
The fracture rate for the large sizes of the 60-degree Shiley (C-C)
valve is now thought to be as much as five times higher than previously
estimated. Valve fracture is often fatal.
FDA believes that the risk of fracture of these large valves, over an
eight-year period, when implanted to substitute for the heart's mitral
valve, may be high enough for doctors and patients to consider replacing
currently intact valves in some individuals.
"When a critical device such as a heart valve is found to have a problem
that could result in death or serious injury, FDA has an obligation to see
that doctors and patients are notified so that they can consider the new
information in deciding on a course of action," said FDA Commissioner David
A. Kessler, M.D.
About 23,000 people in the United States and Canada have 60-degree
-MORE-
Page 2, P92-7, Shiley Valve
Shiley (C-C) valves. These valves were removed from the market in 1986.
FDA has received about 350 reports of 60-degree valve fractures among the
roughly 82,000 C-C valves implanted worldwide.
FDA requested Shiley to notify patients with these valves of the
fracture problems, in a program begun in 1990 after the agency became aware
of some risk of fracture. However, replacement of intact valves was not
recommended at that time because the surgical risk was thought to far
outweigh the risk of fracture. (See press release P90-60, Dec. 11, 1990.)
The risk of fracture depends on the age of the patient, valve size and
valve position. The rate may be as high as 0.8 percent per year for people
under 50.
Which patients should now have their heart valve replaced is a decision
that should be made on a case by case basis by physicians, based on the new
fracture figures and the patient's medical status, lifestyle and wishes.
Since certain patients may be at higher risk for valve fracture than was
previously known, it is important for patients with the larger valves,
especially those in the higher risk group--younger patients with the larger
valves implanted in the mitral position--to talk to their doctors about what
the new risk numbers mean to them, Dr. Kessler said.
FDA asked Shiley to send letters to all patients with C-C heart valves
informing them of the increased risk of fracture. Patients in the affected
groups will be provided with both the new and old fracture risk figures for
various valve sizes and age groups and will be advised to discuss the new
information with their doctors. Patients unaffected by the new figures will
be reassured that the new data do not apply to them.
-MORE-
Page 3, P92-7, Shiley Valve
Shiley was also asked to send a letter to physicians, enclosing a copy
of the study upon which the higher estimates are based and discussing the
option of valve replacement.
"It is important to remember that this new risk information applies only
to Shiley C-C valves, not to other makes and models," said James S. Benson,
director of FDA's Center for Devices and Radiological Health.
"Even with the C-C valves, the increased fracture risk is confined to
certain size valves. The risk figures have not appreciably changed for C-C
valve patients who have the smaller size valves."
The study on which the new estimates of risk are based was published in
the Feb. 1 British medical journal The Lancet.
###