FDA
Enforcement Report
The FDA Enforcement Report is published weekly by the Food and
Drug Administration,
U.S. Public Health Service, Department of Health and Human
Services. It contains information
on actions taken in connection with agency regulatory
activities.
May 12, 1999 99-19
RECALLS AND FIELD CORRECTIONS:
FOODS -- CLASS I
________
PRODUCT
Lascco Seafood Cocktail, in 4 ounce glass jars. (Note:
approximately a year ago the firm changed the name of this
product from Seafood & Crab Cocktail to Seafood Cocktail. Hence
product may be on the market under both names.). Recall #F-372-9.
CODE
Product is not coded but can be identified by UPC code
0-728404-7.
MANUFACTURER
Los Angeles Smoking and Curing Company, Inc., Los Angeles,
California.
RECALLED BY
MANUFACTURER, by telephone on March 24, 1999,
followed by letter, and by press release on March 26, 1999.
Firm-initiated recall ongoing.
DISTRIBUTION
Nevada, California, Washington state, Alaska, Colorado, Idaho,
Texas, Utah, Oklahoma, Hawaii, Arizona, Montana, Oregon.
QUANTITY
4,000 cases (12 jars per case) were distributed.
REASON
Product contains undeclared egg white, snow crab, wheat starch,
and wheat flour.
________
PRODUCT
Lumpia products (Filipino style egg rolls): a) Vegetable Lumpia
(Plain); b) Lumpia Vegetable with Shrimp. Product packaging
contains 14 pieces per bag and 14 bags per case.
Recall #F-375/376-9.
CODE
All product codes produced and distributed between September 22,
1998 and February 8, 1999. This code appears on the packaging of
both products being recalled:
"0009228-0002089".
MANUFACTURER
Fil-Am Specialty Foods, Inc., Los Angeles, California.
RECALLED BY
Manufacturer, by telephone on March 12, 1999, followed by mail
and fax on March 22, 1999, and by press release on April 6, 1999.
Firm-initiated recall complete.
DISTRIBUTION
California.
QUANTITY
a) 143 cases; b) 390 cases were distributed.
REASON
Products contain undeclared egg whites.
RECALLS AND FIELD CORRECTIONS:
FOODS -- CLASS II
________
PRODUCT
Macaroni & Cheese Dinners, in 7.25 ounce boxes, under the
following labels: a) Safeway Mac `N Cheese; Distributed by
Safeway Inc., UPC 21130-50598
b) Food Lion; Distributed by Food Lion, Inc., UPC
35826-00491
c) Best Yet; Packed for Scrivner, Inc., UPC 42187-40712
d) IGA; Distributed by IGA, Inc., UPC 41270-81500
e) Sunny Select; Distributed by Super Store Industries, UPC
17544-10677
f) Super C; Prepared for Briska Inc., UPC 62609-53881
g) Carriage Trade; Prepared for Continental Sales Ltd.,
UPC 56595-24206
h) Glencourt; Prepared for Glencourt Distributors,
UPC 8203-50001. Recall #F-369-9.
CODE
FEB2400C21, FEB2500A21, FEB2500B21, FEB2500C21, FEB2600A21,
FEB2600B21, FEB2600C21, MAR0400B21 and MAR0400C21.
MANUFACTURER
Gilster-Mary Lee Corporation, Steeleville, Illinois.
RECALLED BY
Gilster-Mary Lee Corporation, Chester, Illinois, by press release
on April 9, 1999, and by telephone on April 12, 1999, followed by
fax. Firm-initiated recall ongoing.
DISTRIBUTION
Alabama, California, Colorado, Florida, Mississippi, Nebraska,
North Carolina, Ohio, Pennsylvania, South Carolina, Tennessee,
Utah, Virginia, Wisconsin Canada.
QUANTITY
11,280 cases(48 boxes per case) were distributed; firm estimated
that 50 percent of product remained on market at time of recall
initiation.
REASON
Product may contain pieces of glass.
________
PRODUCT
Ultra Slim Fast Creamy Milk Chocolate Ready to Drink Meal, in 11
ounce steel cans. Recall #F-370-9.
CODE
BEST BY APRIL 2000 ETMC 9053 B02 **:**.
MANUFACTURER
Slim-Fast Nutritional Foods, LLC, Covington, Tennessee.
RECALLED BY
Slim-Fast Foods Company, West Palm Beach, Florida, by letter
faxed on or about April 15, 1999. Firm-initiated recall ongoing.
DISTRIBUTION
Nationwide and Puerto Rico.
QUANTITY
7,613 cases (24 cans per case) were distributed.
REASON
Some cans of product are contaminated with a caustic cleaning
solution.
________
PRODUCT
Periyar Banana Chips, net weight 200 grams.
Recall #F-373-9.
CODE
All lots.
MANUFACTURER
Anns Outdoor Catering, Pala, P.O. Kerala, India.
RECALLED BY
Parel Overseas, Jersey City, New Jersey, by telephone on January
28, 1999. Completed recall resulted from sample analysis and
follow-up by the New York State Department of Agriculture and
Markets.
DISTRIBUTION
New York, Pennsylvania, Texas, Michigan.
QUANTITY
Approximately 40 cases were distributed.
REASON
The product contains undeclared FD&C Yellow No. 5.
______
PRODUCT
Holtermann's Cakes: a) Extra Moist Vanilla Square, packaged in a
19 ounce cardboard box; b) Extra Moist Brides Cake, packaged in a
18 ounce cardboard box.
Recall #F-377/378-9.
CODE
All product coded with "best if sold by" dates on or before OCT 6
1998 (all date codes prior to the initiation of the recall).
MANUFACTURER
Albert Holtermann, Inc., Staten Island, New York.
RECALLED BY
Manufacturer, by visit on October 19, 1998. Completed field
correction (relabeling) resulted from sample analysis and follow-
up by the New York State Department of Agriculture and Markets.
DISTRIBUTION
New York.
QUANTITY
Approximately 100 boxes of each product were distributed.
REASON
a) This cake was found to contain undeclared FD&C Yellow No. 5,
FD&C Yellow No. 6, FD&C Red No. 40, and FD&C Blue
No. 1.
b) This cake was found to contain undeclared FD&C Yellow No. 5,
FD&C Red No. 3, and FD&C Blue No. 1.
________
PRODUCT
Circus Treats brand Tiger Eye Jumbo Gum Balls, 96 pieces per
plastic bag, net weight 960 grams. Product of Egypt.
Recall #F-379-9.
CODE
None.
MANUFACTURER
Unknown. Product of Egypt.
RECALLED BY
Sultana Crackers, Inc., Bronx, New York (importer), by letter on
September 10, 1998. Completed field correction (relabeling)
resulted from sample analysis and follow-up by the New York State
Department of Agriculture and Markets.
DISTRIBUTION
Undetermined.
QUANTITY
Firm estimates none remain on market.
REASON
Product contained an unapproved color additive.
________
PRODUCT
Sun Golden brand Almond Cookies, artificially flavored,
in 1 pound 8 ounce cardboard boxes. Recall #F-380-9.
CODE
None.
MANUFACTURER
Sun Golden Cookies Corporation, Brooklyn, New York.
RECALLED BY
Manufacturer, by telephone and visit on November 9, 1998.
Completed field correction (relabeling) resulted from sample
analysis and follow-up by the New York State Department of
Agriculture and Markets.
DISTRIBUTION
New York.
QUANTITY
Approximately 4,000 boxes were distributed,
REASON
Product contained undeclared FD&C Yellow No. 6.
________
PRODUCT
Breaded frozen fish in poly lined cases:
a) Cod - 2 oz. pieces, in 10 lb cases
b) Halibut pieces - 2 oz. pieces, 3 oz. pieces (V.S.I. label),4
oz. Pieces, 5 oz pieces, in 10 lb. Cases
c) Halibut fillets - 4 oz. pieces in 10 lb. Cases.
d) Halibut patties - 3 oz. Pieces, 4 oz. pieces each in 10 lb
cases.
e) Burgerville sandwich in 20 lb. cases
f) Burgerville Fish-N-Chips in 20 lb. cases.
Recall #F-386/391-9.
CODE
Products a-d: all products bearing date codes prior to Jan 01 99.
Products e-f: all products bearing date codes prior to and
including Apr 08 99.
MANUFACTURER
Vancouver Seafood, Inc., Vancouver, Washington.
RECALLED BY
Manufacturer, by letter on April 22, 1999. Firm-initiated recall
ongoing.
DISTRIBUTION
a-d) Washington state, Oregon, California;
e-f) Oregon.
QUANTITY
a-d) Approximately 3,970 pounds; e-f) Approximately 6,000-8,000
pounds were distributed.
REASON
Products contain undeclared allergens as follows:
a-d contain undeclared FD&C Yellow # 5, FD&C Yellow # 6, and
dehydrated skim milk.
e-f contain undeclared wheat flour, FD&C Yellow 5, FD&C Yellow
#6, and dehydrated skim milk.
RECALLS AND FIELD CORRECTIONS:
FOODS -- CLASS III
________
PRODUCT
Bee's Queen brand Pure Chinese Honey, in 10 ounce glass jars.
Recall #F-371-9.
CODE
None.
MANUFACTURER
Shanghai Native Produce I/E Corporation, Shanghai, China.
RECALLED BY
Tai Wing Hong Importer, Inc., Brooklyn, New York, by letter on
October 30, 1998. Completed recall resulted from sample analysis
and follow-up by the New York State Department of Agriculture and
Markets.
DISTRIBUTION
New York.
QUANTITY
450 cartons (48 jars per carton) were imported; firm estimates
none remains on the market.
REASON
The product, falsely labeled as "Pure Honey", is adulterated as
it was found to contain sugars of cane or corn origin.
________
PRODUCT
Fred Meyer Parmesan Cheese, in 3 ounce cardboard container.
Recall #F-374-9.
CODE
June 02 99 MS2.
MANUFACTURER
Masson Cheese Corporation, Vernon, California.
RECALLED BY
Fred Meyer, Inc., Portland, Oregon, by E-mail on February 25,
1999, and April 2, 1999. Firm-initiated recall complete.
DISTRIBUTION
Alaska, Washington state, Idaho, Montana, Oregon, Utah.
QUANTITY
156 cases (24 units per case) were distributed.
REASON
Product was contaminated with mold.
________
PRODUCT
Bottled flavored water in 20 fluid ounce plastic bottles, labeled
in part, "Natural Choices Pure Water":
a) Strawberry Kiwi Twist; b) Lemon Kiwi Twist;
c) Tangerine Twist. Recall #F-382/384-9.
CODE
Products are labeled with the production date and time.
a) Strawberry Kiwi: 04 29 98 1600 through 04 29 98 1615
b) Lemon Kiwi: 04 30 98 1015 through 04 30 98 1030
c) Tangerine: 04 30 98 1515 through 04 30 98 1530.
MANUFACTURER
Cascade Clear Water, Burlington, Washington.
RECALLED BY
Unique Labeling, Inc., Portland, Oregon, by fax on March 17,
1999. Firm-initiated recall complete.
DISTRIBUTION
Utah.
QUANTITY
The firm estimated that there were 100 cases of each flavored
water with suspect code distributed.
REASON
Product may have been contaminated with mold.
RECALLS AND FIELD CORRECTIONS:
COSMETICS -- CLASS II
________
PRODUCT
Wet'n Wild Mega Lash Mascara, packaged in tube-like containers.
Recall #F-381-9.
CODE
Lot numbers: 161298 and 171298.
MANUFACTURER
AM Cosmetic Products Company, Brooklyn, New York.
RECALLED BY
AM Cosmetic Products Company, North Arlington, New Jersey, by
letter on February 11, 1999. Firm-initiated recall ongoing.
DISTRIBUTION
Nationwide and international.
QUANTITY
28,938 units were distributed.
REASON
Product is contaminated with Burkholderia cepacia.
RECALLS AND FIELD CORRECTIONS:
DRUGS -- CLASS II
________
PRODUCT
Compressed Oxygen USP, Rx in high pressure E and D size
cylinders. Recall #D-208-9.
CODE
Lot #31C992.
MANUFACTURER
Red Ball Oxygen Company, Inc., Shreveport, Louisiana.
RECALLED BY
Manufacturer, by telephone on April 12, 1999. Firm-initiated
recall ongoing.
DISTRIBUTION
Louisiana and Texas.
QUANTITY
38 cylinders were distributed.
REASON
Failure to document the analysis of potency and identity testing.
RECALLS AND FIELD CORRECTIONS:
BIOLOGICS -- CLASS II
________
PRODUCT
Red Blood Cells. Recall #B-733-9.
CODE
Unit #10592-2836.
MANUFACTURER
United Blood Services, Scottsdale, Arizona.
RECALLED BY
Blood Systems, Inc., Scottsdale, Arizona, by letter dated January
16, 1999. Firm-initiated recall complete.
DISTRIBUTION
Arizona.
QUANTITY
1 unit was distributed.
REASON
Blood product was collected from a donor who reported travel to
an area designated as endemic for malaria.
________
PRODUCT
a) Red Blood Cells; b) Platelets; c) Fresh Frozen Plasma;
d) Recovered Plasma. Recall #B-736/739-9.
CODE
Unit numbers:
a) 7436082 7438251 6747771 7438231 7437101 7618378
7437094 7618397 7438239 6747754 7438253 7436081
6747767 7436128 6747772 7437099 7437102 6747759
7436109 7436089 7437095 7437093 7437092 7437104
7437103 7463091 7437096 7436116 7437107
b) 6747754 6747767 7438253 7438239 7438231 6747771
6747772 7438251 6747759;
c) 7436130 7436128 6747759
d) 6747767 7436081.
MANUFACTURER
New York Blood Services, New York, New York.
RECALLED BY
Manufacturer, by letters dated August 19, 1998, followed by
telephone and fax. Firm-initiated recall complete.
DISTRIBUTION
New York and Switzerland.
QUANTITY
a) 29 units; b) 9 unit; c) 3 units; d) 2 units.
REASON
Blood products may have been incorrectly tested for antibody to
the hepatitis C virus encoded antigen (anti-HCV).
RECALLS AND FIELD CORRECTIONS:
BIOLOGICS -- CLASS III
________
PRODUCT
Red Blood Cells. Recall 729-9.
CODE
Unit #16340-3116.
MANUFACTURER
Blood Systems, Inc., Hattiesburg, Mississippi.
RECALLED BY
Blood Systems, Inc., Scottsdale, Arizona, by telephone on
September 24, 1997, and by letter dated October 22, 1997. Firm-
initiated recall complete.
DISTRIBUTION
Mississippi.
QUANTITY
1 unit was distributed.
REASON
Blood products were collected from a donor who had taken the
antibiotic Keflex, two days prior to donation.
________
PRODUCT
a) Red Blood Cells; b) Platelets. Recall #B-731/732-9.
CODE
Unit #16339-7085.
MANUFACTURER
Blood Systems, Inc., Meridian, Mississippi.
RECALLED BY
Blood Systems, Inc., Scottsdale, Arizona, by letter dated October
9, 1997. Firm-initiated recall complete.
DISTRIBUTION
Mississippi.
QUANTITY
1 unit of each component was distributed.
REASON
Blood products were collected from a donor taking the drug
Navane.
________
PRODUCT
Platelets, Pheresis. Recall #B-734-9.
CODE
Unit #36000-1481.
MANUFACTURER
United Blood Services, Ventura, California.
RECALLED BY
Blood Systems, Inc., Scottsdale, Arizona, by letters dated
November 3, 1998, and January 15, 1999. Firm-initiated recall
complete.
DISTRIBUTION
California.
QUANTITY
1 unit was distributed.
REASON
Blood product was collected from a donor who had not been
screened for antibodies and tested for syphilis.
RECALLS AND FIELD CORRECTIONS:
DEVICES -- CLASS II
________
PRODUCT
25 G Whitacre Needle Spinal Tray, intended to be used in the
administration of a regional or local anesthesia to a patient.
The tray contains 5% Lidocaine with 7.5% Dextrose, 2 ml. Recall
#Z-817-9.
CODE
Catalog #4965-25, Lot #903541, EXP August 2000.
MANUFACTURER
Sims Portex, Inc., Keene, New Hampshire.
RECALLED BY
Manufacturer, by telephone on March 26, 1999. Firm-initiated
recall complete.
DISTRIBUTION
Colorado and Indiana.
QUANTITY
3 cases (10 units per case) were distributed.
REASON
Product labeled as sterile was not sterilized.
________
PRODUCT
Hyperinflation Sets and Resuscitator. Recall #Z-831/832-9.
CODE
Lot Numbers:
98287-5025 98288-5062 98288-5084 98288-5085
98288-5087 98288-5089 98288-5090 98288-5091
98288-5093 98289-5094 98292-5118 98292-5123
98293-5128 98295-5222 98299-5251 98299-5260
98306-5404 98306-5419 98306-5420 98309-5463
98313-5475.
MANUFACTURER
Owens-BiGam Medical Company, Newland, North Carolina.
RECALLED BY
Manufacturer, by letter on November 25, 1998. Firm-initiated
recall complete.
DISTRIBUTION
Arizona, California, Colorado, Georgia, Illinois, Kansas,
Louisiana, Maryland, Minnesota, Missouri, New Jersey, North
Carolina, Ohio, Oregon, Pennsylvania, Texas.
QUANTITY
128 cases and 63 samples were distributed.
REASON
Damage to the oxygen tubing could cause leaks in the tubing that
would result in inconsistent flow rates.
________
PRODUCT
Dual Purpose Rocker Instrument, Part #4009, 6.35" size, non-
powered surgical instrument intended to help introduce/seat the
spinal brace rods deeper into the adult hooks or variable locking
screws to allow easier application of the slider or hex nut.
Recall #Z-852-9.
CODE
Lot numbers: 20072 and 20147.
MANUFACTURER
P.A.K Manufacturing, Inc., Melbourne, New Jersey (contract
manufacturer).
RECALLED BY
Interpore Cross International, Irvine, California (responsible
firm), by telephone on February 18 and 22, 1999, and by letter
dated February 18, 1999. Firm-initiated recall complete.
DISTRIBUTION
California, Kentucky, Illinois, Massachusetts, Michigan,
Minnesota, Missouri, North Carolina, Nebraska, Tennessee, Texas,
Virginia.
QUANTITY
27 units were distributed.
REASON
The device will not allow the proper assembly of a slider into a
hook when used in accordance with the product instructions for
use.
________
PRODUCT
a) 5 Fr Angiographic Catheters Model Nos. SPC-454D, SPC-454E, and
SPC-454F; b) 5 Fr Pressure/Volume Catheters; Model Nos. SPC-465A,
SPC-550, SSD-795, and SSD-814.
Recall #Z-853/854-9.
CODE
All lot numbers.
MANUFACTURER
Millar Instruments, Houston, Texas.
RECALLED BY
Manufacturer, by letter on March 19, 1999. Firm-initiated recall
ongoing.
DISTRIBUTION
Nationwide, Canada, Japan.
QUANTITY
290 units were distributed.
REASON
A coagulated clot was found within the catheter which were made
of nylon instead of polyurethane as specified in the firmís
510(k).
________
PRODUCT
Quinton Q-Cath Hemodynamic Analysis System, a hemodynamic
recording system, which acquires, displays, stores and may print
out waveform data acquired from a patient undergoing a cardiac
catheterization procedure:
a) Q-Cath; b) Q-Cath II; c) Q Cath II NT; d) Q-Cath DS;
e) Q-Cath Remote Station. Recall #Z-855/859-9.
CODE
Units with version 4.5 software.
MANUFACTURER
Quinton Instrument Company, Bothell, Washington.
RECALLED BY
Manufacturer, by letter on October 1, 1998, followed by visit.
Firm-initiated field correction ongoing.
DISTRIBUTION
Nationwide, Canada, The Netherlands, Thailand, New Zealand,
Egypt.
QUANTITY
Approximately 81 units were distributed.
REASON
Device has potential to display previous patient's blood pressure
________
PRODUCT
GEM Implantable Cardioverter Defibrillator:
a) Model No. 7227Cx; b) Model No. 7227.
Recall #Z-860/861-9.
CODE
All serial numbers.
MANUFACTURER
Medtronic Med Rel, Inc., Humacao, Puerto Rico; Medtronic,
Switzerland Manufacturing Operations, Tolochenaz, Switzerland.
RECALLED BY
Medtronic, Inc., Minneapolis, Minnesota, by letter on April 2,
1999. In the recall action, a voltage test is being used to
identify defibrillators which have premature battery depletion.
Firm-initiated recall ongoing.
DISTRIBUTION
Nationwide and international.
QUANTITY
Approximately 5,000 devices were distributed.
REASON
A high current drain in some of the defibrillators causes
premature battery depletion.
________
PRODUCT
FlexiMount Ceiling Track Mount System with Trolly, Model
#0102180, part of the mounting system designed to connect the
Burton Outpatient Plus, Outpatient II, Coolspot and Coolspot
lights to the ceiling. Recall #Z-862-9.
CODE
Serial numbers: 940080-940250 and 97253-97313.
MANUFACTURER
Burton Medical Products Corporation, Chatsworth, California.
RECALLED BY
Manufacturer, by letter on April 13, 1999. Firm-initiated recall
ongoing.
DISTRIBUTION
Nationwide and Canada.
QUANTITY
122 units were distributed.
REASON
The device has the potential to fall because setscrews used to
secure the down-tube on the trolley were omitted from the
installation pack.
________
PRODUCT
Pressure Reduction Valve, Part #20132-065, a component of the
Argon Plasma Coagulation (APC) Unit. Recall #Z-863-9.
CODE
Serial numbers: M0005001 to M0005050.
MANUFACTURER
ERBE, Tubingen, Germany,
RECALLED BY
ERBE USA, Inc., Marietta, Georgia, by visit and by letters sent
on April 5, 1999. Firm-initiated recall ongoing.
DISTRIBUTION
Nationwide and Canada.
QUANTITY
50 units with the affected serial numbers have been distributed.
REASON
The valve has the potential to separate from the Argon tank
because it does not fit securely in the tank.
________
PRODUCT
NC-1 Floor to Ceiling Mount, an accessory to a microscope used
for various surgical procedures. Recall #Z-864-9.
CODE
Identification number is KBC 002, with fabrication (serial)
Numbers 10-2044 and lower.
MANUFACTURER
Oberlikon Contraves, A.G., Zurich, Switzerland.
RECALLED BY
Carl Zeiss, Inc., Thornwood, New York, by letter on April 1,
1999. Firm-initiated field correction ongoing.
DISTRIBUTION
Florida, Kentucky, Texas, North Carolina, Illinois, Alabama,
Missouri, Ohio, Louisiana, Arizona, Georgia, California, Hawaii,
Delaware, Colorado, Utah, Pennsylvania, Arkansas.
QUANTITY
52 units were distributed.
REASON
The device has the potential to fall during use.
________
PRODUCT
Storz Silicone Tire Implant for Retinal Detachment Surgery
a) Catalog No. E5381 721 (287 Style);
b) Catalog No. E5381 620 (78g Style). Recall #Z-867/868-9.
CODE
Lot numbers: a) MI70870, MI70880, MA83090, MC80850, MC80860,
MC80870, and MC80880; b) MA81790.
MANUFACTURER
Vesta, Inc., Franklin, Wisconsin.
RECALLED BY
Bausch and Lomb Surgical, Inc., St. Louis, Missouri, by letter
dated February 24, 1999 and the week of April 19, 1999. Firm-
initiated recall ongoing.
DISTRIBUTION
Nationwide and international.
QUANTITY
473 pouches were distributed.
REASON
The package labeling may not correctly identify the size and
style of the contents.
________
PRODUCT
Bayer ADVIA 120 Hematology System, a quantitative automated
hematology analyzer that provides a leukocyte differential count
and reticulocyte analysis for in-vitro-diagnostic use in clinical
laboratories:
a) ADVIA 120 Hematology System, Catalog No. 453-0024-03;
b) ADVIA 120 Hematology System, Catalog No. 453-0024-04.
Recall #Z-869/870-9.
CODE
All serial numbers.
MANUFACTURER
Bayer Diagnostic Manufacturing Ltd., Swords, Co Dublin, Ireland.
RECALLED BY
Bayer Corporation, Elkhart, Indiana, by customer bulletin sent on
February 9, 1999. Firm-initiated recall ongoing.
DISTRIBUTION
Nationwide and international.
QUANTITY
Approximately 920 units were distributed.
REASON
The device may give erroneous results because of the valve gasket
deterioration in the Baso channel check valve, and the software
designed will not flag significant discrepancies in the results
between the two methodologies at extremely low counts.
________
PRODUCT
Ocufilcon D 55 Visibility Tinted Contact Lenses .
Recall #Z-871-9.
CODE
Product Names (trade names)Involved In Recall:
Brand (-)(+): Product Code: Brand: Product Code:
Aqualens Disp. 55 5133 Proflex 55 2210
Aqualens Disp. 55 5134 Proflex 55 2222
Aquatech 55 2225 Prosite 55 2252
Aquatech 55 2226 Prosite 55 2253
Clinasoft 55 2202 Sofmed 55 2228
Clinasoft 55 2214 Softech 55 2248
Diagnostic 55 2263 Softech 55 2249
Diagnostics 55 2262 Softique 55 2235
Hydroflex 55 2264 Softmed 55 2227
Hydron Biomedics 55 2208 Ultraflex 7/14 55 2201
Hydron Biomedics 55 2220 Ultraflex 7/14 55 2213
Hydrovue 55 2246 Veraflex Eyerned 55 5141
Hydrovue 55 2247 Versaflex 55 2250
Mediflex 2203 Versaflex 55 2251
Mediflex 55 2215 Versaflex Eyerned 55 5143
Optiflex 2209 Optiflex 55 2221
Optiform 55 2254 Optiform 55 2255
P Label Disp. 55 2223 Polysoft 55 2256
Polysoft 55 2257 Private Label 55 2211
Procon 55 2230 Procon 55 2229.
MANUFACTURER
Ocular Sciences, Inc., South San Francisco, California.
RECALLED BY
Manufacturer, by letter on April 26, 1999. Firm-initiated recall
ongoing.
DISTRIBUTION
Nationwide and international.
QUANTITY
1,283,472 lenses were distributed.
REASON
Incomplete seal could compromise product sterility.
________
PRODUCT
Colleague 3 Volumetric Infusion Pump (Triple Channel) for
continuous or intermittent delivery through clinically acceptable
routes of administration such as intravenous, intra-arterial,
subcutaneous, epidural or irrigation of fluid spaces
applications. Recall #Z-872-9.
CODE
Product Code P2M8153, All serial numbers.
MANUFACTURER
Baxter Healthcare PTE. Ltd. Singapore.
RECALLED BY
Baxter Healthcare Corporation, IV Systems Division, Round Lake,
Illinois, by letter April 16, 1999. Firm-initiated field
correction ongoing.
DISTRIBUTION
Nationwide.
QUANTITY
6,992 pumps were distributed.
REASON
A software communication error occurs predominately when all
three channels are in use, causing an alarm condition
(533:320:717:000) that will stop the function of all of the
channels in use. The pump will alarm appropriately, alerting the
user of the failure state.
________
PRODUCT
DMR2 Disposable Manual Resuscitators:
1. Model No. ARXL; 2. Model No. ARML;
3. Model No. ARMLDP; 4. Model No. ARPML;
5. Model No. ARMS; 6. Model No. ARXB;
7. Model No. ARMB; 8. Model No. ARPMB;
9. Model No. ARMBDP; 10. Model No. AVXL;
11. Model No. AVML; 12. Model No. AVPML;
13. Model No. AVMLDP; 14. Model No. AVMS;
15. Model No. AVXB; 16. Model No. AVMB;
17. Model No. AVPMB; 18. Model No. AVMBDP;
19. Model No. CRPML; 20. Model No. CRPMS;
21. Model No. CRPXB; 22. Model No. CRPMB;
23. Model No. CVPML; 24. Model No. CVPMS;
25. Model No. CVPXB; 26. Model No. CVPMB;
27. Model No. CVPMBDP; 28. Model No. IRPML;
29. Model No. IRPXB; 30. Model No. IRPMB;
31. Model No. IVPMS; 32. Model No. IVPXB;
33. Model No. IVPMB; 34. Model No. IVPMBDP.
Recall #Z-874/907-9.
CODE
All lots.
MANUFACTURER
Nellcor Puritan Bennett de Mexico S.A. de C.V., Tijuana, Mexico.
RECALLED BY
Mallinckrodt, Inc., St. Louis, Missouri, by letter on March 30,
1999. Firm-initiated recall ongoing.
DISTRIBUTION
Nationwide and international.
QUANTITY
Approximately 100,716 units were distributed.
REASON
The non-rebreathing valve (NRV) housing assembly may crack
resulting in the swivel elbow fitting becoming loose and
disengaging from the NRV during use.
RECALLS AND FIELD CORRECTIONS:
DEVICES -- CLASS III
________
PRODUCT
Medline Gripper Sport and Gripper II Latex Free Operating Room
Shoe Covers: a) Catalog No. NON27752;
b) Catalog No. NON27758; c) Catalog No. NON27758R;
d) Catalog No. NON27759; e) Catalog No. NON27759R;
f) Catalog No. NON27852; g) Catalog No. NON27144;
h) Catalog No. NON27144XL. Recall #Z-844/851-9.
CODE
Lot Numbers: 307892JC, 362750JC, 376467JC, 376495JC, 393363JC,
393346JC.
MANUFACTURER
Hangzhou Jinchen Knitting & Textiles Co., Ltd., Hangzhou, China.
RECALLED BY
Medline Industries, Inc., Mundelein, Illinois, by undated letter
on April 1, 1999. Firm-initiated recall ongoing.
DISTRIBUTION
Nationwide.
QUANTITY
1,950 cases were distributed.
REASON
The latex free gripper pattern on the shoe covers is made of a
water-soluble substance and may smear and may be more prone to
slippage if exposed to water.
________
PRODUCT
a) Minnova Pelvic Floor Stimulation System;
b) InnoSence Pelvic Floor Stimulation and Electromyography
System. Both systems use mild electrical stimulation to help
control urinary incontinence. Recall #Z-865/866-9.
CODE
Various serial numbers.
MANUFACTURER
EMPI, Inc., St. Paul, Minnesota.
RECALLED BY
Manufacturer, by letter dated March 26, 1999. Firm-initiated
recall ongoing.
DISTRIBUTION
Nationwide and international.
QUANTITY
a) 126 units; b) 73 units were distributed.
REASON
Electrical short at a pin connector may change the location of
the stimulation to an incorrect location.
________
PRODUCT
IMX HCG Calibrators, for in vitro diagnostic use. The IMx hCG
Assay is a Microparticle Enzyme Immunoassay (MEIA) for the
quantitative and qualitative determination of Human Chorionic
Gonadotropin (hCG) in human serum and plasma for the early
detention of pregnancy. Recall #Z-873-9.
CODE
Lot #3A63-01, List No. 47128Q100 EXP 8/9/99.
MANUFACTURER
Abbott Health Products, Inc., Barceloneta, Puerto Rico.
RECALLED BY
Manufacturer, by telephone and letter dated April 20, 1999.
Firm-initiated recall ongoing.
DISTRIBUTION
Nationwide and international.
QUANTITY
210 kits were distributed.
REASON
Decreased F calibrator rates may result in elevated control and
patient values.
________
PRODUCT
Proclear Daily Wear Contact Lens. Recall #Z-908-9.
CODE
Lot B814117.
MANUFACTURER
Biocompatibles Eyecare, Norfolk, Virginia.
RECALLED BY
Manufacturer, by telephone on April 14, 1999. Firm-initiated
recall ongoing.
DISTRIBUTION
Nationwide, Canada, United Kingdom.
QUANTITY
119 lenses were distributed.
REASON
Contact lenses optical power was mislabeled on product.
RECALLS AND FIELD CORRECTIONS:
VETERINARY PRODUCTS -- CLASS II
________
PRODUCT
RUMA Rich Black label Hi-Energy 20 Instant Calf Replacer
(chlortetracycline 20 grams/ton), in 50 pound bags.
Recall #V-067-9.
CODE
Lot numbers: 8L1407, 8J2707, 9B0904.
MANUFACTURER
Strauss Feeds, Watertown, Wisconsin.
RECALLED BY
Manufacturer, by telephone on March 8, 1999. Ongoing recall
resulted from sample analysis and follow-up by the Wisconsin
Department of Agriculture, Trade & Consumer Protection.
DISTRIBUTION
Iowa, South Dakota, Wisconsin.
QUANTITY
600 bags were distributed.
REASON
Product is superpotent - Chlortetracycline is four times the
labeled amount. Product was labeled to contain 20 gm/ton,
analysis found 84 grams/ton.
_______
PRODUCT
Edge Mix Finisher #1 with CSP (Chlortetracycline 100 g/ton,
Sulfathiazole 0.011% and Penicillin 50 g/ton custom mix bulk
medicated swine pre-starter. Recall #V-068-9.
CODE
Bulk feed with manufacture date of 4/2/99.
MANUFACTURER
Cargill, Inc., Flora, Illinois.
RECALLED BY
Manufacturer, by telephone on April 2, 1999. Firm-initiated
recall complete.
DISTRIBUTION
Indiana.
QUANTITY
6,020 pounds were distributed.
REASON
The medicated feed is manufactured with a CSP Type A medicated
article and a CSP Type B medicated feed pack. On 4/2/99 the mill
used the wrong type B medicated feed packs which contained BMD
(acitracin Methylene Disalycylate) instead of CSP. Another feed
was delivered at the same time - Edge Mix Finisher #1 with
fenbendazole. The Edge Mix Finisher #1 with CSP was
inadvertently dumped into the same bin as the fenbendazole feed.
________
PRODUCT
Nutrena Ringmaster TM Show Hog Grow/Fin (BMD- Bacitracin
Methylene Disalycylate) Medicated, in 50 pound bags, used for
increased rate of weight gain and improved feed efficiency.
Recall #V-069-9.
CODE
Lot STK275 - Formula 2888-A.
MANUFACTURER
Cargill Animal Nutrition Divison/Nutrena Feeds, Stockton,
California.
RECALLED BY
Manufacturer, by telephone on January 6, 1999. Firm-initiated
recall complete.
DISTRIBUTION
California.
QUANTITY
943 bags were distributed.
REASON
Product is subpotent. Product contained 19.2% of the labeled
BMD.
MEDICAL DEVICE SAFETY ALERTS:
________
PRODUCT
Techstar XL and Prostarr XL Precutaneous Vascular Surgical
Device, intended for the percutaneous deployment and tying of
sutures for surgical closure of arterial puncture sites resulting
from diagnostic and interventional procedure:
Models: TXL-431-06; TXLS-432-06; TXL-431-07; PXL-431-08;
PXL-431-10. Safety Alert #N-005/009-9.
CODE
All codes.
MANUFACTURER
Perclose, Inc., Redwood City, California.
ALERTED BY
Manufacturer, by safety alert letters of April 28, 1999
describing the labeling concerns and providing additional sticker
labeling for placement on the outer packages.
DISTRIBUTION
Nationwide and international.
QUANTITY
23,960 units were distributed.
REASON
Misinterpretation of sterility labeling on the inner tray by
users may result in contamination and inappropriate use of the
trays after the outer pouch sterility barrier has been broken.
END OF ENFORCEMENT REPORT FOR MAY 12, 1999.
####