10/09/1996

 ENFORCEMENT REPORT FOR 10/09/96 

October 9, 1996                                      96-41

RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS I ============
_______________
PRODUCT        Chocolate Chip Cookie Dough Ice Cream, under
               the following labels: Galliker's -- 3 gallon
               bulk containers, 1/2 gallon containers,
               Leichtman's -- 1/2 gallon round containers,
               CR's Friendly Markets - 1 pint rounds, 1/2
               gallon rounds.  Recall #F-737-6.
CODE           All codes.
MANUFACTURER   Galliker Dairy Company, Johnstown,
               Pennsylvania.
RECALLED BY    Manufacturer, by memorandum on July 12, 1996,
               and by press release on July 26, 1996.  Firm-
               initiated recall complete.
DISTRIBUTION   Pennsylvania, Maryland, Virginia, West
               Virginia.
QUANTITY       Firm estimated that little if any product
               remained on market.
REASON         Product contains undeclared eggs.

_______________
PRODUCT        Jack & Jill Chocolate Chip Cookie Dough Ice
               Cream, in 1/2 gallon containers.  
               Recall #F-738-6.
CODE           1/2 gallons code #23269-10027, dated APR497,
               MAY0997, MAY2997, JUN0597, JUN1397, JUL1097
               (printed on the end flaps of carton).  Pint
               cartons code #23269-18027 dated BB51597 and
               BB71697 (printed on the bottom of cartons). 
MANUFACTURER   Simco Manufacturing, Inc., Cuyahoga Falls,
               Ohio. 
RECALLED BY    Jack & Jill Ice Cream, Bensalem, Pennsylvania,
               by press release on August 6, 1996.  Firm-
               initiated recall complete.
DISTRIBUTION   Pennsylvania, New Jersey, Delaware, Maryland,
               Virginia, West Virginia, New York, District of
               Columbia.
QUANTITY       Undetermined.
REASON         Product contains undeclared eggs.

_______________
PRODUCT        Reinhold's Chocolate Chip Cookie Dough Ice
               Cream, in 1/2 gallon square, 1 quart round,
               and 3 gallon round cardboard bulk containers. 
               Recall #F-739-6.
CODE           All codes.
MANUFACTURER   Reinhold Ice Cream Company, Pittsburgh,
               Pennsylvania.
RECALLED BY    Manufacturer, by press release on July 23,
               1996.  Firm-initiated recall complete.
DISTRIBUTION   Pennsylvania.
QUANTITY       Firm estimated that 70 half-gallon squares, 70
               1-pint rounds and 65 3-gallon bulk containers
               remained on market at time of recall
               initiation.
REASON         Product contains undeclared eggs.

_______________
PRODUCT        Hartley's Choice Chocolate Chip Cookie Dough
               Ice Cream, in 1/2 gallon round cardboard
               containers, distributed by Kings Family
               Restaurants, Inc.  Recall #F-740-6.
CODE           All codes.
MANUFACTURER   King's Country Shoppes, Inc., McKeesport,
               Pennsylvania (distributor).
RECALLED BY    Distributor, by press release on July 24,
               1996.  Firm-initiated recall ongoing.
DISTRIBUTION   Pennsylvania.
QUANTITY       Firm estimated that 108 half-gallon containers
               remained on market at time of recall
               initiation.
REASON         Product contains undeclared eggs.


                             -2-RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS II ===========
_______________
PRODUCT        Eggstra Pasteurized Homogenized Liquid Egg
               Product in cardboard cartons.
               Recall #F-735-6.
CODE           Lot #234-61 USE BY 9-17-96 and Lot 235-61 USE
               BY 9 18 96.
MANUFACTURER   National Food Corporation, Pacific Egg
               Products Division, Arlington, Washington.
RECALLED BY    Manufacturer, by telephone beginning August
               28, 1996.  Firm-initiated recall ongoing.
DISTRIBUTION   Washington state.
QUANTITY       215 cases (30 units per case) of lot 233-61
               and 114 (30 units per case) of lot 234-61 were
               distributed.
REASON         Product is contaminated with salmonella.

_______________
PRODUCT        Lay's Regular Potato chips, in 6 ounce size
               bags.  Recall #F-741-6.
CODE           Manufacturing code 4351A63, Lot No. October
               15, 1996.
MANUFACTURER   Frito-Lay, Inc., Orlando, Florida.
RECALLED BY    Frito-Lay, Inc., Plano, Texas, by computer on
               September 13, 1996.  Firm-initiated recall
               ongoing.
DISTRIBUTION   Florida and Georgia.
QUANTITY       All product were distributed.
REASON         Product contains undeclared dairy ingredients
               and bears an inappropriate identity statement.


RECALLS AND FIELD CORRECTIONS:  FOODS -- CLASS III ==========
_______________
PRODUCT        Apple & Eve 100% Pure & Natural Apple Juice,
               8.45 fluid ounce 3-pak brick pack juice boxes. 
               Recall #F-742-6.
CODE           EXP OCT 30 96B
MANUFACTURER   Golden Town Apple Products, Montreal, Canada.
RECALLED BY    Apple & Eve, Roslyn, New York, by letter June
               13, 1996.  Completed recall resulted from
               sample analysis and followup by New York State
               Department of Agriculture and Markets.
DISTRIBUTION   New York, Vermont, New Jersey.
QUANTITY       1,440 cases were distributed; firm estimates
               none remains on the market.
REASON         Product contains undeclared potassium sorbate,
               a preservative.

_______________
PRODUCT        Large Gumdrops and Spearmint Leaves candy
               packaged in 7 ounce clear plastic bags.  
               Recall #F-743/744-6.
                             -3-CODE           All codes.
MANUFACTURER   Farley Foods USA, Chicago, Illinois.
RECALLED BY    Salem's old Fashioned Candies, Inc., Salem,
               Massachusetts, by telephone.  Completed recall
               resulted from sample analysis and followup by
               State of New York Department of Agriculture
               and Markets.
DISTRIBUTION   Nationwide.
QUANTITY       Approximately 84 packages of gumdrops and 396
               packages of spearmint leaves.
REASON         Product contains undeclared FD&C Yellow No. 5.


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS II ===========
_______________
PRODUCT        Bacitracin Zinc (500 units) and Polymyxin B
               Sulfate (10,000 units) Ophthalmic Ointment
               USP, under the following labels:  Bausch &
               Lomb, Rugby, Medical Ophthalmics, H.L. Moore,
               and Akorn.  Recall #D-275-6.
CODE           (Bausch & Lomb Label) #'s 733661, 683082,
               643531, 593411; (Rugby Labs label, Medical
               Ophthalmics label, and H.L. Moore label) Lot #
               683081; (Akorn label) Lot # 643532.
MANUFACTURER   Bausch & Lomb Pharmaceutical Division, Tampa,
               Florida.
RECALLED BY    Manufacturer, by letter on August 5, 1996.
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       93,491 units were distributed; firm estimated
               that 2,000 units remained on market at time of
               recall initiation.
REASON         Potency is not assured through its shelf-life
               due to possible degradation of bacitracin
               zinc.

_______________
PRODUCT        OTC Dental Cream (toothpaste) labeled as
               manufactured by Colgate-Palmolive, Bombay,
               India.  Recall #D-278-6.
CODE           Undetermined.
MANUFACTURER   Undetermined except for labeled declaration
               which indicated Colgate-Palmolive in Bombay,
               India.
RECALLED BY    International Marketing and Management
               Services, Beverly Hills, California, by letter
               on August 21, 1996, followed by telephone. 
               Firm-initiated recall complete.
DISTRIBUTION   California and the United Kingdom.
QUANTITY       743 cases (144 tubes per case) were
               distributed.

                             -4-REASON         Colgate-Palmolive states that it is
               counterfeit and its use is associated with
               four complaints of mouth irritation.  

_______________
PRODUCT        Oxygen, USP, in size E cylinders.  
               Recall #D-280-6.
CODE           Lot #090996 which included 3 cylinders with
               serial numbers E12699, 28804, and E15539.
MANUFACTURER   Respiratory Home Care Consultants, Inc., West
               Middlesex, Pennsylvania.
RECALLED BY    Manufacturer, by telephone on or about
               September 12, 1996.  Firm-initiated recall
               complete.
DISTRIBUTION   Pennsylvania.
QUANTITY       3 cylinders were distributed.
REASON         Purity was not tested.

_______________
PRODUCT        Apothe'Cure brand Adrenal Cortex Extract
               Injectable, 2 mg/ml, Rx, various formulations
               and vial sizes:
               Adrenal Cortical Extract, 3 ml, 2 mg/ml in
               Sesame Oil, Preservative Free, Sterile
               Injection
               Adrenal Cortical Extract, 10 ml, 2 mg/ml in
               Sesame Oil, Preservative Free, Sterile
               Injection
               Adrenal Cortical Extract, 10 ml, 2 mg/ml in
               Normal Saline, Preservative Free, Sterile
               Injection
               Adrenal Cortical Extract, 30 ml, 2 mg/ml in
               Normal Saline, Preservative Free, Sterile
               Injection
               Adrenal Cortical Extract, 50 ml, 2 mg/ml in
               Normal Saline, Preservative Free, Sterile
               Injection
               Adrenal Cortical Extract, 100 ml, 2 mg/ml in
               Normal Saline, Preservative Free, Sterile
               Injection
               Adrenal Cortical Extract, 3 ml, 2 mg/ml in
               pyrogen free Sterile Water, Preservative Free,
               Sterile Injection
               Adrenal Cortical Extract, 30 ml, 2 mg/ml in
               pyrogen free Sterile Water, Preservative free,
               Sterile Injection.  Recall #D-001/003-7. 
CODE           All codes.
MANUFACTURER   Apothe'Cure, Inc., Dallas, Texas.
RECALLED BY    Manufacturer, by letter dated August 15, 1996. 
               Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       Approximately 3,525 units were distributed.

                             -5-REASON         Product is an unapproved new drug; FDA
               inspection found current good manufacturing
               practice deficiencies and FDA analysis found
               units lacking the labeled active ingredient. 


RECALLS AND FIELD CORRECTIONS:  DRUGS -- CLASS III ==========
_______________
PRODUCT        Royce brand Pindolol Tablets, USP, Rx in 100
               tablet bottles, 10 mg and 5 mg strengths, used
               in the management of hypertension.  
               Recall #D-276/277-6.
CODE           Lot numbers:  MD-1196 (5 mg); MD-1193 (10 mg).
MANUFACTURER   Royce Laboratories, Miami, Florida.
RECALLED BY    Manufacturer, by survey conducted in April 18
               through 29, 1996.  Firm-initiated recall
               complete.
DISTRIBUTION   California, New Jersey, Ohio, Mississippi,
               Michigan, New York, Illinois, Florida,
               Louisiana, Kentucky, Arkansas.
QUANTITY       4,688 10-mg bottles and 4,437 5-mg bottles
               were distributed.
REASON         Normally white tablets may turn pink.

_______________
PRODUCT        Compressed Oxygen USP in aluminum cylinders. 
               Recall #D-279-6.
CODE           Lot numbers: 021396-1    021396-2    022096-1
               022296-1     022296-2    022696-3    022796-1 
               022996-1     030596-1    030596-2    031496-1 
               040496-1     041196-1    041196-2    041696-1
               041696-2     042396-1    042396-2    043096-1
               050996-1     051096-1    051696-1    052196-1
               052196-2     060496-1    060496-2    061196-1
               070296-1     080796-1    080896-1    082996-1
               082996-2     052296-2    091796-1. 
MANUFACTURER   Laurel Medical Supplies, Inc., Ebensburg,
               Pennsylvania.
RECALLED BY    Manufacturer, by visit beginning on September
               24, 1996.  Firm-initiated recall ongoing.
DISTRIBUTION   Pennsylvania.
QUANTITY       Approximately 100 cylinders per month were
               distributed between February 1996 through
               September 1996.
REASON         Purity was below specification or not tested.


_______________
PRODUCT        Nitro-Bid IV (nitroglycerin injection USP),
               packaged in 1-ml., 5-ml., and 10-ml. vials,
               for the treatment of perioperative 

                             -6-               hypertension, for control of congestive heart
               failure in the setting of acute myocardial
               infarction, for the treatment of angina
               pectoris, and for induction of intraoperative
               hypotension.  Recall #D-004-7.
CODES          S00141, S00117, S00116, S00115, S00109,
               S00098, S00082.
MANUFACTURER   Sanofi Winthrop Pharmaceuticals, McPherson,
               Kansas (contract manufacturer).
RECALLED BY    Hoechst Marion Roussel, Inc., Kansas City,
               Missouri (responsible firm), by letter dated
               September 10, 1996.  Firm-initiated field
               correction ongoing.
DISTRIBUTION   Nationwide.
QUANTITY       10,893/1-ml vials between 9/30/94-8/28/96;
               2,584/5-ml vials between 10/23/95-8/12/96;
               2,327/10-ml vials between 6/1/96-4/23/96 were
               distributed.
REASON         Insert incorrectly states the dosage
               concentration as mg/ml and mg/min rather than 
               mcg/ml and mcg/min.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS II =======
_______________
PRODUCT        Platelets.  Recall #B-649-6.
CODE           Unit #29135-0871.
MANUFACTURER   United Blood Services, Chicago, Illinois.
RECALLED BY    Manufacturer, by letter dated June 13, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   Illinois.
QUANTITY       1 unit was distributed.
REASON         Blood product was prepared from a unit of
               Whole Blood designated as a "difficult"
               collection.

_______________
PRODUCT        (a) Red Blood Cells; (b) Recovered Plasma.
               Recall #B-651/652-6.
CODE           Unit #18022-6877.
MANUFACTURER   United Blood Services, Cheyenne, Wyoming.
RECALLED BY    Blood Systems, Inc., Scottsdale, Arizona, by
               letter dated April 1, 1996.  Firm-initiated
               recall complete.
DISTRIBUTION   Tennessee, North Carolina.
QUANTITY       1 unit of each component were distributed.
REASON         Blood products were collected from a donor
               with a history of hepatitis.

_______________
PRODUCT        Blood Grouping Reagent Anti-D (anti-Rho). 
               Recall #B-653-6.

                             -7-CODE           Lot numbers:  D145-1A, IC, ID, IE, IG, IBA,
               IBB, IFA, IFB EXP 12/23/95; D146-1A, IC, ID,
               IBA, IBB EXP 9/15/96; D147-1A, 1B, 1C, 1DA,
               IDB, 1DC EXP 7/5/97.
MANUFACTURER   Dade International of Puerto Rico, Aguada,
               Puerto Rico.
RECALLED BY    Manufacturer, by letter dated December 1995,
               and by fax.  Firm-initiated recall ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       Approximately 11,046 packages were distributed
               between March 22, 1994 and September 22, 1995.
REASON         Blood grouping reagent gave weaker than
               expected antigen typing results.

_______________
PRODUCT        Red Blood Cells.  Recall #B-654-6.
CODE           Unit #0474644.
MANUFACTURER   Coffee Memorial Blood Center, Inc., Amarillo,
               Texas.
RECALLED BY    Manufacturer, by telephone and letters dated
               August 14, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION   Texas.
QUANTITY       1 unit.
REASON         Blood products were collected from a donor
               diagnosed with a history of malignant prostate
               cancer.

_______________
PRODUCT        Aplitest, Tuberculin, Purified Protein
               Derivative Multiple Puncture Device for
               Intradermal Testing.  Recall #B-656-6.
CODE           Lot #00504P EXP 9/95.
MANUFACTURER   Parke-Davis, Division of Warner-Lambert
               Company, Rochester, Michigan.
RECALLED BY    Manufacturer, by letter on March 27, 1995. 
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       1,848 units were distributed.
REASON         Product failed antigen potency test.

_______________
PRODUCT        Red Blood Cells.  Recall #B-660-6.
CODE           Unit #137088.
MANUFACTURER   Blood Bank of Alaska, Anchorage, Alaska.
RECALLED BY    Manufacturer, by letter dated July 13, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   Alaska.
QUANTITY       1 unit was distributed.
REASON         Blood product was collected from a donor who
               traveled to a malarial endemic area.

                             -8-_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c)
               Recovered Plasma.  Recall #B-661/663-6.
CODE           Unit #685832.
MANUFACTURER   Community Blood Center, Dayton, Ohio.
RECALLED BY    Manufacturer, by letters dated January 24,
               1996, and February 9, 1996.  Firm-initiated
               recall complete.
DISTRIBUTION   Ohio and California.
QUANTITY       1 unit of each component.
REASON         Blood products were collected from a donor
               with a reported history of a contaminated
               needle stick.

_______________
PRODUCT        (a) Red Blood Cells; (b) Recovered Plasma.
               Recall #B-670/671-6.
CODE           Unit #BB141067.
MANUFACTURER   Blood Bank of Alaska, Anchorage, Alaska.
RECALLED BY    Manufacturer, by letter dated August 22 and 
               23, 1995.  Firm-initiated recall complete.
DISTRIBUTION   Alaska and California.
QUANTITY       1 unit of each component were distributed.
REASON         Blood products were collected from a donor
               with a history of ear piercing within 12
               months of donation.

_______________
PRODUCT        Red Blood Cells.  Recall #B-676-6.
CODE           Unit numbers:  2N8887, 4H1726, 1S3086, 1S3082.
MANUFACTURER   Puget Sound Blood Center, Seattle, Washington.
RECALLED BY    Manufacturer, by telephone on October 5, 1995. 
               Firm-initiated recall complete.
DISTRIBUTION   Washington state.
QUANTITY       4 units were distributed.
REASON         Red Blood Cells were labeled incorrectly as
               Red Blood Cells, Leukocytes Removed by
               Filtration.

_______________
PRODUCT        Recovered Plasma.  Recall #B-679-6.
CODE           Unit #23-15112.
MANUFACTURER   University Medical Center-Mesabi, Hibbing,
               Minnesota.
RECALLED BY    Manufacturer, by telephone on April 19, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   Pennsylvania.
QUANTITY       1 unit was distributed.
REASON         Blood product tested repeatedly reactive for
               the antibody to the human immunodeficiency
               virus types 1 and 2 (anti-HIV-1/2), Western
               blot indeterminate.

                             -9-_______________
PRODUCT        (a) Platelets; (b) Fresh Frozen Plasma.
               Recall #B-680/681-6.
CODE           Unit #KX70309.
MANUFACTURER   Central Blood Bank, Pittsburgh, Pennsylvania.
RECALLED BY    Manufacturer, by letter dated July 7, 1994. 
               Firm-initiated recall complete.
DISTRIBUTION   Pennsylvania.
QUANTITY       1 unit of each component were distributed.
REASON         Blood products containing anti-K antibodies
               were labeled as negative for the unexpected
               antibodies. 

_______________
PRODUCT        Red Blood Cells.  Recall #B-688-6.
CODE           Unit numbers:  10FQ01278, 10FH13013,
               10FG14964.
MANUFACTURER   American Red Cross Blood Services,
               Springfield, Missouri.
RECALLED BY    Manufacturer, by telephone on October 10,
               1995.  Firm-initiated recall complete.
DISTRIBUTION   Missouri.
QUANTITY       3 units were distributed.
REASON         Blood products were collected from a donor who
               traveled to an area designated as endemic for
               malaria.

_______________
PRODUCT        Red Blood Cells.  Recall #B-691-6.
CODE           Unit #R00541.
MANUFACTURER   Tacoma-Pierce County Blood Bank, doing
               business as Cascade Regional Blood Services,
               Tacoma, Washington.
RECALLED BY    Manufacturer, by fax on January 12, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   Washington state. 
QUANTITY       1 unit.
REASON         Blood product was collected from a donor with
               a history of cancer.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets.
               Recall #B-693/694-6.
CODE           Unit #9520956.
MANUFACTURER   Lorain County Blood Bank, Elyria, Ohio.
RECALLED BY    Manufacturer, by telephone on December 11,
               1995, followed by letter dated January 16,
               1996.  Firm-initiated recall complete.
DISTRIBUTION   Ohio.
QUANTITY       1 unit of each component was distributed.

                            -10-REASON         Blood products were collected from a donor who
               traveled to an area designated as endemic for
               malaria.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c) Fresh
               Frozen Plasma; (d) Recovered Plasma.  
               Recall #B-699/702-6.
CODE           Unit numbers:  (a) 94-0191, 94-0833, 94-2032,
               94-2174, 94-2453, 94-3196, 94-3767, 94-3970,
               94-4047, 95-0770, 95-1138, 95-1726, 95-2450,
               95-3084, 95-3125, 95-3903, 96-0623, 96-0975,
               96-1428; 
               (b) 94-3196, 95-1726, 95-3125, 96-0623;
               (c) 94-4047, 95-1138, 96-0975, 96-1428;
               (d) 94-0191, 94-0833, 94-2032, 94-2174, 
               94-2453, 94-3196, 94-3970, 95-0770, 95-1726,
               95-2450, 95-3084, 95-3125, 95-3903, 96-0623.
MANUFACTURER   Merced County Blood Bank, Merced, California.
RECALLED BY    Manufacturer, by letter dated July 9, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   California and New York.
QUANTITY       (a) 19 units; (b) 4 units; (c) 4 units (d) 14
               units were distributed.
REASON         Blood products tested non-reactive to anti-
               HIV-1, were collected from donors who
               previously tested repeatedly reactive for
               anti-HIV-1, EIA negative, Western Blot
               negative and donors were subsequently re-
               entered improperly.

_______________
PRODUCT        (a) Red Blood Cells; (b) Recovered Plasma. 
               Recall #B-704/705-6.
CODE           Unit #GJ51025.
MANUFACTURER   Blood Bank of San Bernardino and Riverside
               Counties, San Bernardino, California.
RECALLED BY    Manufacturer, by letter dated February 28,
               1994.  Firm-initiated recall complete.
DISTRIBUTION   California.
QUANTITY       1 unit of each component was distributed.
REASON         Blood products were collected from a donor
               whose spouse has history of IV drug use.


RECALLS AND FIELD CORRECTIONS:  BIOLOGICS -- CLASS III ======
_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c)
               Recovered Plasma.  Recall #B-646/648-6.
CODE           Unit #5Y2587.
MANUFACTURER   Puget Sound Blood Center, Seattle, Washington.

                            -11-RECALLED BY    Manufacturer, by letters dated July 27, 1995
               and August 11, 1995.  Firm-initiated recall
               complete.
DISTRIBUTION   Washington state, California.
QUANTITY       1 unit of each component was distributed.
REASON         Blood products were collected from a donor who
               received a corneal transplant within one year
               of donation.

_______________
PRODUCT        Whole Blood.  Recall #B-650-6.
CODE           Unit 29132-7664.
MANUFACTURER   United Blood Services, Chicago, Illinois.
RECALLED BY    Manufacturer, by telephone on February 24,
               1996.  Firm-initiated recall complete.
DISTRIBUTION   Illinois.
QUANTITY       1 unit was distributed.
REASON         Autologous blood product tested positive for
               the hepatitis B surface antigen (HBsAg).

_______________
PRODUCT        Recovered Plasma.  Recall #B-655-6.
CODE           Unit #0474644.
MANUFACTURER   Coffee Memorial Blood Center, Inc., Amarillo,
               Texas.
RECALLED BY    Manufacturer, by letter on August 14, 1995. 
               Firm-initiated recall complete.
DISTRIBUTION   California.
QUANTITY       1 unit.
REASON         Blood products were collected from a donor
               diagnosed with a history of malignant prostate
               cancer.

_______________
PRODUCT        (a) Red Blood Cells; (b) Platelets; (c)
               Recovered Plasma.  Recall #B-657/659-6.
CODE           Unit numbers:  (a) 0557005 EXP 12/10/94; (b)
               0557005 EXP 11/3/94; (c) 0557005 EXP 12/13/94.
MANUFACTURER   Coffee Memorial Blood Center, Inc., Amarillo,
               Texas.
RECALLED BY    Manufacturer, by telephone on January 4, 1995. 
Firm-initiated recall complete.
DISTRIBUTION   Tennessee and Texas.
QUANTITY       1 unit of each component were distributed.
REASON         Blood products were collected from a donor who
               received a tissue transplant (surgery) within
               a year of donation.

_______________
PRODUCT        Source Plasma.  Recall #B-669-6.
CODE           Unit #XQ64057.
MANUFACTURER   Community Bio-Resources, Inc., Birmingham,
               Alabama.
                            -12-RECALLED BY    Manufacturer, by letter dated October 13,
               1995.  Firm-initiated recall complete.
DISTRIBUTION   Austria.
QUANTITY       1 unit was distributed.
REASON         Blood product was untested for syphilis.

_______________
PRODUCT        Red Blood Cells.  Recall #B-677-6.
CODE           Unit #07H25891.
MANUFACTURER   American Red Cross Blood Services, Tucson,
               Arizona.
RECALLED BY    Manufacturer, by telephone on February 21,
               1992.  Firm-initiated recall complete.
DISTRIBUTION   Arizona.
QUANTITY       1 unit was distributed.
REASON         Blood product was labeled with an extended
               expiration date.

_______________
PRODUCT        Irradiated Red Blood Cells.  Recall #B-678-6.
CODE           Unit numbers:  C52846, E65455, C52958, C52212.
MANUFACTURER   Tri-Counties Blood Bank, Inc., Santa Barbara,
               California.
RECALLED BY    Manufacturer, by telephone on March 17, 1994. 
               Firm-initiated recall complete.
DISTRIBUTION   California.
QUANTITY       4 units were distributed.
REASON         Irradiated blood products were labeled with an
               extended expiration date.

_______________
PRODUCT        Red Blood Cells.  Recall #B-689-6.
CODE           Unit numbers:  0275565 and 5521402.
MANUFACTURER   Inland  Northwest Blood Center, Spokane,
               Washington.
RECALLED BY    Manufacturer, by telephone in September 1995. 
               Firm-initiated recall complete.
DISTRIBUTION   Washington state.
QUANTITY       2 units were distributed.
REASON         Blood products were labeled with an extended
               expiration date.

_______________
PRODUCT        Red Blood Cells.  Recall #B-690-6.
CODE           Unit numbers:  2241462 and 2241422.
MANUFACTURER   Inland Northwest Blood Center, Spokane,
               Washington.
RECALLED BY    Manufacturer, by letter dated June 12, 1995. 
               Firm-initiated recall complete.
DISTRIBUTION   Washington state.
QUANTITY       2 units were distributed.
REASON         Blood products were labeled with an extended
               expiration date.
                            -13-_______________
PRODUCT        Recovered Plasma.  Recall #B-692-6.
CODE           Unit #R00541.
MANUFACTURER   Tacoma-Pierce County Blood Bank, doing
               business as Cascade Regional Blood Services,
               Tacoma, Washington.
RECALLED BY    Manufacturer, by fax on January 12, 1996. 
               Firm-initiated recall complete.
DISTRIBUTION   California.
QUANTITY       1 unit.
REASON         Blood product was collected from a donor with
               a history of cancer.

_______________
PRODUCT        Red Blood Cells.  Recall #B-703-6.
CODE           Unit numbers:  07F71730 and 07F71731.
MANUFACTURER   American Red Cross, Tucson, Arizona.
RECALLED BY    Manufacturer, by telephone on or about June
               23, 1993.  Firm-initiated recall complete.
DISTRIBUTION   Arizona.
QUANTITY       2 units were distributed.
REASON         Blood products were exposed to unacceptable
               temperatures.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS II =========
_______________
PRODUCT        Hand-Held Controllers for the Maxxus Dual Head
               Imaging Systems and the Optima Integrated
               Multi-Dector Imaging System.  
               Recall #Z-1235/1236-6.
CODE           All serial numbers.
MANUFACTURER   IGE (International General Electric) Ltd., St.
               Albans, United Kingdom.
RECALLED BY    General Electric Company, Medical Systems
               Division, Waukesha, Wisconsin, by field
               modification instructions issued on July 24,
               1996.  Firm-initiated field correction
               ongoing.
DISTRIBUTION   Nationwide and international.
QUANTITY       315 units were distributed.
REASON         Excessive electronic noise in the hand-held
               controllers could cause unintended motion of
               the gantry, which could result in collision of
               the gantry with the patient.

_______________
PRODUCT        C Optima Imaging System, Model 9SYS0088 with
               Granite Central Processing Units (CPU's), a
               nuclear medicine imaging system.  
               Recall #Z-1237-6.
CODE           Systems with granite CPU's.

                            -14-MANUFACTURER   General Electric Company, Medical Systems
               Division, Waukesha, Wisconsin.
RECALLED BY    Manufacturer, by field modification
               instructions issued on August 30, 1995.  Firm-
               initiated field correction complete.
DISTRIBUTION   Florida, Georgia, North Carolina, Tennessee,
               Singapore.
QUANTITY       5 units were distributed.
REASON         There could be unintended motion of the
               gantry.

_______________
PRODUCT        Threaded Plastic Cannula Sleeves:
               (a) Catalog No. 140834;
               (b) Catalog No. 140835;
               (c) Catalog No. 140836;
               (d) Catalog No. 140916.  
               Recall #Z-1238/1241-6.
CODE           All lot numbers.
MANUFACTURER   Gebrueder Pajunk, Geisingen, Germany.
RECALLED BY    Pilling Weck, Inc., Research Triangle Park,
               North Carolina, by letter on November 1, 1995. 
               Firm-initiated recall complete.
DISTRIBUTION   Nationwide and international.
QUANTITY       64 sleeves were distributed; firm estimates
               none remains on the market.
REASON         Capacitive coupling may occur when the plastic
               cannula is used in conjunction with a metal
               cannula.


RECALLS AND FIELD CORRECTIONS:  DEVICES -- CLASS III ========
_______________
PRODUCT        One Day ACUVUE (etafilcon A) Daily Wear
               Disposable Contact Lenses, -2.75 power/9.0 mm
               base curve.  Recall #Z-1242-6.
CODE           Lot Numbers:  1601200868, 1601200967,
               1601201063, 1601201162.
MANUFACTURER   Vistakon, Inc., Johnson and Johnson Vision
               Products, Inc., Jacksonville, Florida.
RECALLED BY    Manufacturer, by telephone from September 5 to
               10, 1996.  Firm-initiated recall complete.
DISTRIBUTION   Nationwide.
QUANTITY       1,751 unit cartons (30 units per carton) were
               distributed.
REASON         The correctly labeled outside carton of the
               9.0mm base curve product may be labeled with
               an incorrect base curve (8.5mm) on the inner
               primary package label.

                            -15-